Re: tracking cell division

From: Alice L. Givan (Alice.L.Givan@dartmouth.edu)
Date: Wed Jul 10 2002 - 16:00:33 EST


We have used both PKH26 and CFSE.  Both are, with our cells, toxic at high concentrations
and both seem less or not toxic at low concentrations.  The trick,  if you want to track
many divisions,  is that you want to start out with cells that are as bright as possible.
So the titration window between getting bright cells and not killing them is critical.
The scary thing we have found,  particularly with CFSE, is that at somewhat high
concentrations you inhibit some reactions (like antigen specific proliferation) but
not other reactions (like ConA stimulated proliferation).  So you do need to test with
tritiated thymidine uptake under your own cell culture conditions  to make sure that
cell division is not affected by the CFSE or PKH labeling.

>From first principles,  it seems to me that it might be safer to use PKH dyes rather
than CFSE because the CFSE will label proteins and may affect the function of some
of them.  Since the PKH dyes only involve lipids,  the chance of specific toxicity
might not be as great.  But I know that many people have great success with CFSE ---
and it certainly is much cheaper than the PKH dyes.

Alice

Alice L. Givan
Englert Cell Analysis Laboratory
of the Norris Cotton Cancer Center
Dartmouth Medical School
Lebanon, New Hampshire NH 03756
tel 603-650-7661
fax 603-650-6130
givan@dartmouth.edu



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:26:15 EST