RE: Auto-fluorescence

From: Jeff Barry (JBarry@picr.man.ac.uk)
Date: Fri Jun 07 2002 - 02:39:37 EST


Iaj,

We have found that if you acquire cells on a LogFL1(GFP) vs LogFL2 dotplot
and set the instrument settings, so that the cells fall in the first decade,
you can discriminate gfp +ve cells from background fluorescence by applying
some FL2-FL1 compensation. This should bring any gfp +ves down towards the
FL1 axis. If you keep the PMT settings for F1 and FL2 roughly equal you
probably find you may need to apply about 15% compensation. This works well
with cells but I can't say for protoplasts. Good luck.

Regards,

Jeff Barry
Flow Cytometry
Paterson Institute for Cancer Research
Manchester
UK