Dear flow-ers, Thanks a lot for your help. We have checked all potential sources for contamination and/or impurity in the fluidic system. Eventually we have resolved the problem. Our APC beads tube was contaminated/unclear (instead of using one drop we just pipete 15 mkl with a tip and we have possibly contaminated the tube) and now we just are using another lot. Best regards, Stojan Central Laboratory of Immunology, National Center of Infectious and Parasitic Diseases 1504 Sofia, 26, Yanko Sakazov Blvd, Bulgaria --- Stojan Dimitrov <stoianski_fl@yahoo.com> wrote: > > Dear flowers, > > In our laboratory we have a problem with > disinfection > of the fluidic system of our Flow Cytometer. We use > a > self made sheath fluid which contains azid as a > substance to stop the grouth of bacteria. > Unfortunetly > when we calibrate the cytometer with “Calibrite > beads” > there are many events above the treshold of 300 on > FSC. Is there any substance which we might add to > the > sheath fluid to stop the contamination of the > fluidic > system and which would be harmless to the optics and > silicon tubes. > Any help will be appreciated. Thank you in advance > > Best regards, > Stoian Dimitrov > > Central Laboratory of Immunology, National Center of > Infectious and Parasitic Diseases 1504 Sofia, 26, > Yanko Sakazov Blvd, Bulgaria > > > > __________________________________________________ > Do You Yahoo!? > Try FREE Yahoo! Mail - the world's greatest free > email! > http://mail.yahoo.com/ __________________________________________________ Do You Yahoo!? Yahoo! Movies - coverage of the 74th Academy Awards® http://movies.yahoo.com/
This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:26:04 EST