Flowers, Re: calculating the volume of a drop: A few discrete experts have patted me on the back for finally beginning to figure it out. I will pass the word on to all of you. If you collect the sheath flowing from the stream for, say, 2 minutes, all you need to do is divide that volume by the total number of drops generated in those 2 minutes. And that gives you the volume of a drop. You don't even need to have the drop drive on. Easy peasy (as my kids used to say). Alice Alice L. Givan Englert Cell Analysis Laboratory of the Norris Cotton Cancer Center Dartmouth Medical School Lebanon, New Hampshire NH 03756 tel 603-650-7661 fax 603-650-6130 givan@dartmouth.edu
This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:26:02 EST