I've had quite a lot of questions and comments about red-excited
Alexa fluorochromes since my response to the APC conjugation
question, so I thought I'd make a general post about the questions
and summarize some of the comments.  The most common is in regard to
what filters we use for 633 and 660 detection.  It's not entirely
clear to me at this point what is best for 660, since the filters
that best match the emission spectra do not give the best
fluorescence detection.  We're still trying to empirically determine
the best filter, but right now I would suggest something around 695
to 700 bp 15 to 20.  Standard APC filters (eg, 660) work pretty well
for A633, although we're still tweaking this as well.  It's worth
mentioning that things may be a little different depending on which
dichroics/beam splitters are in front, but the above are pretty good
starting points.

In regard to Marck Schlomchik's comments about Alexa 647, there's no
question that the absorption spectra for this fluor are the closest
match for HeNe excitation.  However, if your goal is two red laser
fluorescences, it's going to be pretty tough compensating 647 from
633 or 660.  Richard Haugland correctly points out that A680 might be
an option for a red combination with A647, since it has a pretty nice
secondary peak around HeNe excitation wavelengths.  In our case,
however, we're looking to simplify things by generating conjugate
pairs that can be used on all our instruments.  The limitation here
is our modified BD LSR (which by the way is a fantastic instrument,
and I have no commercial interest in BD), which by virtue of the
compact and extremely complex lightpaths won't allow us to use 647
and 680 together (at least not without ordering a number of custom
filters and dichroics; maybe if BD or Omega or Chroma wants to help
us out we can...).  So for us, using two red fluors excited by a HeNe
with detection compatibility on all our instruments, 633/660 works
and works quite well.
--