Re: HO33342 into 670/20 filter???

From: Larry Arnold (lwarma@med.unc.edu)
Date: Fri Jan 11 2002 - 19:37:14 EST


Maris

As long as you have the lasers together you can not resolve the problem
except by compensation (this is the problem in the Coulter machines because
even thought the lasers can be spatially resolved the emissions from
separate lasers are not focused through pinholes).  If you separate the
lasers and use a Vantage or MoFlo you will solve the problem (mostly on the
MoFlo probably completely on the Vantage) without a need for compensation
because the emission from each laser is focused through spatially resolved
pinholes and thus in theory the HO emission will not get to the APC
detector.  For the most part this works in spite of worries about light
piping.  The emission from HO33342 at this wavelength is only a small
percentage of the total HO emission, however, since the total emission from
the HO is very large the HO670 emission relative to an antibody-APC stain
(ie. the HO is very bright)can be a problem.  This is  a much bigger
problem in the FITC channels since the HO emission at this wavelength is
much more than at the 670nm range.  Some people think you should be able to
solve the problem with filters but that only displays a flaw in their
thinking since there is really HO emission across a wide wavelength range.

Larry

At 03:15 PM 1/11/2002, you wrote:
>          Reply to:   Re: HO33342 into 670/20 filter???
>Hi Larry,
>Thanks for your help.  HO33342 does emit at that wavelength?  That would
>explain it.  It was very easy to compensate (nothing extreme), so I'll
>stop worrying about it, and go forward with the experiments.
>We ran the experiment on a FACS Vantage (so the UV and 647 excitation were
>coaxial).  Any suggestions?  Next week we will try it (again) on the
>Vantage, and on a Mo-Flo as well to see if changing the instrument makes a
>difference.
>
>Thanks again,
>Maris
>Larry Arnold wrote:
> >Maris
> >
> >What instrument are you using?  HO33342 does emit significantly at
> that >wavelength so you can get cross-talk between lasers.  After I know
> what >instrument you are using I may be able to suggest something.
> >
> >Larry
> >
> >At 11:12 AM 1/10/2002, you wrote:
> >
> >>Hi Everyone,
> >>Happy New Year to All!
> >>
> >>I am at a loss to explain why cells that are stained with Hoechst33342
> are >>increasing
> >>the background in my APC signal...HELP!  We were looking at DNA
> using >>HO33342 (excited by
> >>UV, emission through a 450/20 filter) and APC (surface receptor,
> using >>647nm excitation
> >>and emission through a 670/20 filter).  There was nothing out of
> the >>ordinary visible
> >>during the alignment (no laser noise increasing the background of
> either >>signal, etc).
> >>We set the APC negative control within the first decade, looked at
> the >>APC+ control, and
> >>saw a distinct positive population.  When we put on a sample that
> was >>stained with HO
> >>(APC-), the signal in the APC channel went into the higher end of
> the >>second decade.
> >>At first, we were splitting the two signals with a 510LP, and
> after >>witnessing the
> >>shift in background, we switched to a 640LP...we still saw the same shift.
> >>We compensated the HO out of the APC signal, and the experiment seemed
> to >>work as
> >>expected (read: the researcher was happy with his results), but
> does >>anyone have any
> >>idea what happened?  I've been looking at every spectrum figure I
> can >>find, and have
> >>looked at the filter specs and I can't find any reason for
> that >>shift.  Any advice,
> >>as always, is sincerely appreciated. Thanks,
> >>Maris
> >>
> >>Maris A. Handley
> >>Dana-Farber Cancer Institute
> >>J415
> >>44 Binney St.
> >>Boston, MA 02115
> >>(617)632-3139
> >>maris_handley@dfci.harvard.edu
> >
> >Larry W. Arnold, Ph.D.
> >Associate Professor
> >Director, Flow Cytometry Facility
> >Department of Microbiology and Immunology
> >Lineberger Comprehensive Cancer Center
> >CB# 7290
> >University of North Carolina at Chapel Hill
> >Chapel Hill, NC 27599
> >Phone: 919-966-1530
> >FAX: 919-962-8103
> >
> >RFC822 header
> >-----------------------------------
> >
> > Received: from europa.dfci.harvard.edu ([155.52.44.41])
> by >PHSEXCHICI2.Partners.org with SMTP (Microsoft Exchange Internet Mail
> Service Version >5.5.2650.21)
> >       id CT5XLRV9; Fri, 11 Jan 2002 15:57:18 -0500
> > Received: from redtail.med.unc.edu (redtail.med.unc.edu [152.19.4.7])
> >       by europa.dfci.harvard.edu (8.10.2+Sun/8.11.3) with ESMTP id
> g0BKtq600444
> >       for <Maris_Handley@dfci.harvard.edu>; Fri, 11 Jan 2002 15:55:52
> -0500 (EST)
> > Received: from PC231H602A.med.unc.edu (pc231h602a.med.unc.edu
> [152.19.43.97])
> >       by redtail.med.unc.edu (8.9.3/8.9.3) with ESMTP id PAA19243;
> >       Fri, 11 Jan 2002 15:57:00 -0500 (EST)
> > Message-Id: <5.0.2.1.2.20020111161015.020dd388@imap-ns.med.unc.edu>
> > X-Sender: lwarma@imap-ns.med.unc.edu
> > X-Mailer: QUALCOMM Windows Eudora Version 5.0.2
> > Date: Fri, 11 Jan 2002 16:12:20 -0500
> > To: Maris Handley <Maris_Handley@dfci.harvard.edu>,
> >    Cytometry Mailing List <cytometry@flowcyt.cyto.purdue.edu>
> > From: Larry Arnold <lwarma@med.unc.edu>
> > Subject: Re: HO33342 into 670/20 filter???
> > In-Reply-To: <200201101711.MAA18384@flowcyt.cyto.purdue.edu>
> > Mime-Version: 1.0
> > Content-Type: text/plain; charset="us-ascii"; format=flowed
> >

Larry W. Arnold, Ph.D.
Associate Professor
Director, Flow Cytometry Facility
Department of Microbiology and Immunology
Lineberger Comprehensive Cancer Center
CB# 7290
University of North Carolina at Chapel Hill
Chapel Hill, NC 27599
Phone: 919-966-1530
FAX: 919-962-8103



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:25:55 EST