Re: CFSE

From: Alice L. Givan (Alice.L.Givan@dartmouth.edu)
Date: Mon Nov 26 2001 - 09:39:56 EST


I  need to continue the discussion and to respond to the very valid points made by
Andrew Wells.  Unlike Andrew, we do have some  good data showing CFSE-inhibiton
of cell division (as viewed by tritiated thymidine uptake) -- with some specific
antigen activators  (but not with ConA as a polyclonal stimulator).  This occurred at 5
micromolar CFSE concentration.   I have also seen some good data from other labs using
CFSE at lower concentrations where inhibition did not occur even with specific antigens.
So I believe that CFSE concentration is an important factor (as would be expected). In
our experiments, the CFSE labelling was done before cell activation  (although I do take
Andrew's point that labeling activated cells that have already made their "important"
proteins may be an important additional difficulty that needs to be considered when
using CFSE).  We have been using human cells. I have some feeling that human cells
may differ from mouse cells .


I agree completely that CFSE,  in many (but not all) protocols, gives much tighter
distributions than do the PKH dyes.  And therefore one can often do precursor frequency
and proliferation calculations on CFSE-labeled cells without needing to resort to
ModFit-type deconvolution of the histograms (although deconvolution might help in the
frequent cases where even the CFSE histogram distributions from the dividing cells are
not completely distinct).  This deconvolution is always required with PKH-labeling and
has the problems (especially related to non-linear log amps) that Andrew described.
So,  with respect to tight labelling,  CFSE is much better than the PKH dyes. Again,
there may be some human/mouse differences here.  The work showing lovely CFSE peaks
during cell culture is often in mouse systems.  I could be wrong on this --- and would
like to hear from people using CFSE labelling of human T-cells.

Unlike Andrew, we find that PKH-labelling  is very stable --- and we routinely
look at cells cultured over 8 days.  So,  I don't feel that the stability of PKH
labeling is a problem in our protocols.  And the intensity of PKH labelling is fine
(especially since we cannot go very high with the CFSE staining,  without worrying
about functional inhibition).  We usually have to back off on the brightness of our
PKH labeling so that we don't have compensation problems.

One of the main factors that Andrew did not mention is that CFSE is much cheaper than the
PKH dyes.  So,  my general opinion (with which Andrew might agree!), is that you should
use the CFSE dyes as long as you have confirmed (with, for example, tritiated thymidine
uptake assays) that CFSE does not inhibit cell proliferation with your cells under your
staining conditions and with your .  So the message is that, with either dye,  you need
to be careful that the staining is not affecting the very function that you are studying.

Alice

Alice L. Givan
Englert Cell Analysis Laboratory
of the Norris Cotton Cancer Center
Dartmouth Medical School
Lebanon, New Hampshire NH 03756
tel 603-650-7661
fax 603-650-6130
givan@dartmouth.edu



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:40 EST