Re: EMA/FITC/PE

From: Mario Roederer (roederer@drmr.com)
Date: Fri Sep 21 2001 - 19:29:15 EST


Ray, EMA has a very similar emission spectrum to propidium iodide.
It can be easily distinguished and even properly compensated from PE
as well as fluorescein (and 9 other colors as well!).  EMA excites
well at 488, and can be read in your "FL3" channel--the same one you
use for Cy5PE or PerCP.

Make sure to do a compensation sample for the EMA.  Typically, we
will kill cells using a 50C shock for a few minutes in order to
generate a good positive control sample.

The EMA staining protocol is available in the Current Protocols for
Cytometry; we can send you our current one if you wish (it's quite
similar).

mr



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:32 EST