Re: Vybrant Cell Labelling Solution - Aggregates Help?

From: Cissy M. Geigerman (cmgeiger@uncg.edu)
Date: Tue Sep 04 2001 - 12:25:31 EST


Peter,

I have had a similar situation.  For a monocyte adhesion
study, I was trying to load human U937 or THP-1 monocytes,
with 7-amino-4-chloromethylcoumarin (CMAC)from Molecular
Probes, and the cells all stuck to the walls of the
staining vessel. This stain also is incompatible with serum
in the media.  Serum-less media may make the cells more
likely to stick. I was able to solve this problem by making
sure the cells were shaking or rotating during the 30 min
loading period.

Hope this helps.

Cissy Geigerman
University of North Carolina at Greensboro

On Fri, 31 Aug 2001 13:49:07 +0100 Peter Kierulf
<peter.kierulf@ioks.uio.no> wrote:

>
>
> Dear Flowers
>
> Always some problems (probably for many of you).
>
> For some time we have tried to label human monocytes (cryopreserved and
> thawed) with Molecular Probes Vybrant Cell-Labeling Solution. Our major
> problem is aggregation of the cells.
>
> We have followed the protocol carefully, using cells as recomended of 1 x
> 106 /mL, without serum, and adding 5 mL of the Cell-Labeling Solution per
> mL of cell suspension, incubating for varying times (30 minutes 37oC,
> spinning the cells and washing a couple of times.
>
> Generally the cells have been aggregating or adhering  to our plastic tubes
> (Falcon Polypropylene tubes 352070).
>
> Our intention has been to study cell migration in TransWell systems using
> this labeling technique.
>
> All suggestions to circumvent aggregation is most welcome.
>
> Thanks!
>
>
> Peter Kierulf
>
> Peter Kierulf,MD PhD. Professor I
> Clinical Chemistry Dep.
> Ulleval University Hospital
> N-0407 OSLO, NORWAY
>
> Telephone: +47 22 11 94 86
> Telefax:   +47 22 11 81 89
> _-----------------------------------Mobil telphone: 922.85111
> E-mail (RFC Formate): peter.kierulf@ioks.uio.no
> E-mai. (SA Formate): /G=peter/S=kierulf/OU=ioks/O=uio/PRMD=uninett/C=no

----------------------
Cissy Geigerman
Dept.of Nutrition and Food Service Systems
Univ. of North Carolina at Greensboro
(336)334-5313



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:30 EST