Re: Spinnaker Laser Problem

From: Arnold Pizzey (a.pizzey@UCL.ac.uk)
Date: Fri Aug 24 2001 - 02:39:47 EST


Hello there Claudio,

It could well be that your plasma tube is overpressure: if the tube has not
been run for a time the gas fill circuit may have given it a shot of gas at
startup and overcompensated: as the tube warms up the tube goes
overpressure and shuts off, at the moment, the bore pressure after warm-up
is *just* below the cutoff value. with luck, running the laser will
eventually bring the bore pressure back down, if it does not, and you
experience repeated cutouts  the solution is (if I remember correctly from
our I90) to disable the overpressure cutout (which is what I suspect is
closing down the tube) and run the laser for an extended period (anything
up to a few weeks!) to bring the bore pressure down. There should be a test
point on the laser control PCB which will give you a bore pressure readout
(there will be a conversion factor from mV to millitorr) If the
manufacturers are friendly they should be able to give you this info
-certainly Coherent in the UK were very helpful in getting my overpressure
Innova-90 running again)
A good on line resource for lasers is http://www.cord.org/lev3.cfm/48

Arnold.


>At 11:37 AM 8/20/01 +0200, Claudio Vallan wrote:
>
>>Hi,
>>
>>I am experiencing some problems with our Spinnaker Laser (from
>>Spectra-Physics) and would like to ask if somebody can tell me what is
>>happening. The Spinnaker is a water cooled 1 Watt Ion Laser. It's maximal
>>output is 272 mW as measured on the multimeter of the remote controller. I
>>usually use it at 30 mW, where it consumes 7.8 AMP.
>>
>>After coming back from Holidays (Laser has been 3 Weeks without running)
>>the laser started to give troubles. It would turn up, but after a certain
>>period at 30 mW it just would stop lasing (more than 30 minutes). The same
>>effect could be reached in a few seconds if turning the power higher than
>>150 mW. Though, it  did not completely stop leasing, as there were sparks
>>of blue light generated. The indicators on the remote control did not show
>>anything unusual even when the laser was sparking (AC ON green / Emission
>>white/ Sum Fault, Interlock and Water Flow OK). The only unusual thing
>>which I had noticed was that some weeks ago, when putting the laser at
>>maximum output it took it some seconds to reach 272 mW while usually that
>>is not the case. After I let it run for a 30 Minutes at full output this
>>behaviour had disappeared.  I do not think the problem I have has to do
>>with cooling as there seem to be no problem with the water (Though I can
>>not be 100% sure).
>>
>>BD told me that I have to buy a new laser and while searching money to do
>>that I replaced the Spinnaker with the second Laser (I am working on a
>>Vantage), so we can at least work with 3 colours in the meanwhile. But when
>>I tried to turn on the Spinnaker the next day (just for fun) the laser was
>>working normal again. On the third day I had again the problems, but just
>>once, when turning it on a second time it just worked normal. Since then it
>>seems to have no problems.
>>
>>Can somebody tell me what happens? Will I need a new laser very soon? Could
>>the laser be repaired? Does Spectra-Physics still exist and what is their
>>address? Is there a good book out there about Lasers with the topics: how
>>they work, how you maintain, troubleshoot and repair them?
>>
>>Answers to any of above questions very appreciated
>>
>>cheers
>>
>>Claudio
>>
>>===================================================
>>Claudio Vallan PhD                          Phone Lab:      031 / 632 88 76
>>FACS-LAB DKF                               Phone Office:  031 / 632 99 68
>>University of Bern                         E-Mail:  vallan@dkf7.unibe.ch
>>c/o Institute of Pathology
>>Murtenstrasse 31                          Insel hosptial area only:
>>3010 Bern                                     Beeper:           181 67 59
>>Switzerland
>>===================================================
>
>Derek Schulze
>Flow Cytometry and Confocal Microscopy Facility Manager
>Cancer Research Labs
>Queen's University
>Kingston, ON
>Canada
>
>
>

_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/
	Arnold Richard Pizzey
	Department of Haematology
	Royal Free and University College London Medical School
	98 Chenies Mews
	London WC1E 6HX
	U.K

	voice:	+44 020-7679-6234
	Fax:	+44 020-7679-6222
	email:	a.pizzey@ucl.ac.uk
_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:29 EST