Re: 3 color fluorescense on a Scope?

From: David Coder (dcoder1@qwest.net)
Date: Sat Aug 11 2001 - 00:04:29 EST


Jim,

All three at the same time will be a problem--especially if you use
Cychrome--since there will be lots of fluorescent overlap between the
FITC/PE pair and the PE/Cychrome pair.

Best choice would be a combination of fluorochromes matched to specific
excitation/emission filters in the filter cube, and (if you want to get
pictures--highly recommended since memory can fade faster than FITC in the
bright summer sun) a high resolution digital camera.

That is FITC, PE, Texas Red would be good choices with the appropriate
filters they give you single excitation/emission images--one for each
fluorochrome--with no fluorescence overlap. That means, FITC would excite
well with a 490 bandpass and emit around 520nm; PE would excite well around
515nm and emit around 575nm; and Texas Red (or the equivalent Alexa dye)
would excite well around 568nm and emit well around 620nm. All should be
doable with a Hg arc lamp, a Xe arc lamp, or even a W/halogen. So, you find
a field of interest, compose the image, take three separate images (without
moving slide), and do the various combinations of overlays in something like
Photoshop. (N.B.: The absolute fluorescence emission--and their subsequent
combined overlays--will have little meaning in terms of relative antigen
content unless you use a proper calibration standard. And that's another
topic altogether.)

Dave
----------------
David M. Coder, Ph.D.
email: dcoder1@qwest.net
tel./messages: 206-499-3446


----- Original Message -----
From: "jim phillips" jphillip@med.miami.edu
To: cyto-inbox
Sent: Friday, August 10, 2001 9:33 AM
Subject: 3 color fluorescense on a Scope?


>
> Hello to everyone,
> This is a little off the subject of flow... but I have an investigator
> that want to do 3 color cell surface fluorescense on a microscope.  The
> question is .. what three fluorochomes can she use..  The microscope is
> just a normal fluorescense scope.  I know FITC and PE but is there a
> third?? Can you use Cychrome???   She wants to view all three at the
> same time..  any help would be appreciated.. Thanks...
> Jim Phillips
> University of Miami School of Medicine.
>
> ps.... I just finished my music CD .. entitled ... "222.... 2 old?  2
> fat? 2 ugly?"  You can listen to it on mp3..
>
>  www.mp3.com/jim_phillips
>
>