Re: Titrating Isotype control Antibodies

From: Mark Kukuruga (kukuru@med.umich.edu)
Date: Thu Jul 12 2001 - 13:42:02 EST


Jeannine,
You'll want to relate the isotype to its companion(s) . . . so, whatever concentration
you choose for the specific antibodies, compare that to an equal concentration of the
matching isotype.  This is best related in equivalent amounts of protein, not simply
a dilution factor, so you do need to assay for content.
It's also important (many may say most important) to compare fluorochrome/protein
ratio of controls and specific antibodies.  Differences in conjugation efficiency will
dramatically effect the eventual fluorescence, and will also effect "noise" detected
with an isotype.  This has been one of the many arguments against the use of isotypes
in general . . . since the F/P ratio is often different, using these as controls for
establishing +/- cutoffs may be misleading.  Personally, I believe isotypes are useful
in most cases in the absence of better negatives, and if one accounts for protein
content and F/P differences.

I hope this doesn't prompt a new "isotype vs. no isotype" debate . . .
MAK.

--