Hi Danni, It could be a range of things, mostly user-fixable (eg broken lead to sheath level detector, jammed or broken level sensor, faulty pressure transducer, water in the air line, faulty low laser power, poor sample seal, amongst others), what does it say in your status control panel? On my Calibur on run with a tube on, the sample voltage (inversely related to pressure) stabilises at around 6.4 on low flow rate and 6.1 on high (these voltage are higher than for my FACSort), if the voltage on your status panel works this way then you can rule out the sample seal (bal seal?). The laser power should be around 15mW shortly after switch on (it might drop into standby, where the power stays at around 3.5 - 5 mW after a while depending on the cause of your problem), if it's significantly lower than 15mW it might prevent the machine switching to ready, you can either locate the potentiometer that controls laser power and turn it up, or seek further advice. If the current is high you'll probably need a new tube or head eventually. The waste and sheath should show up as "OK", if not look further for connection problems. In the case of the sheath tank, you could try removing the probe from the tank and inverting the probe so that the float slides up the shank, it should register OK when inverted, and empty when upright (if it never shows "OK", try short circuiting the connector using a paper clip until you get a new probe). If short-circuiting doesn't work, investigate the wiring under the slide-out tray. If the waste shows as full when it's not, check for short circuits (unplugging the probe from the machine should register "OK") If, on run, with the sample-support arm swung to one side, the sample voltage doesn't rise to 10.23 or nearby, it could be that the air pump or regulators need attention, or it could be that water has been introduced into the air line from the sample port - follow the tubing inside the machine and replace the hydrophobic filters (0.2 micron PTFE syringe filters) Lastly, it could be that you've got an air-lock in the sheath filter, try bleeding it into the waste tank by releasing the tube clamp. This isn't an exhaustive list, nor are all of these suggestions necessarily relevant to the symptom you've reported - if you send me some feedback I'll try to be more specific if you don't get better-informed advice elsewhere (B-D or Ray Lannigan perhaps? - how about putting the troubleshooting flowchart on the web Ray?) Good Luck, Ray Ray Hicks ________________________________________________________________________ |University of Cambridge |Tel 01223 330149 | |Department of Medicine |Fax 01223 336846 | |Level 5, Addenbrookes Hospital |e-mail <rh208@cus.cam.ac.uk> | |Hills Road Cambridge |Web http://facsmac.med.cam.ac.uk | |CB2 | http://www.med.cam.ac.uk | |UK |ftp server ftp://131.111.80.78 | |_________________________________|_____________________________________| > From: "Dhanwanthie Ramduth" <Ramduthd1@nu.ac.za> > Date: Fri, 11 May 2001 16:34:48 +0200 > To: Cytometry Mailing List <cytometry@flowcyt.cyto.purdue.edu> > Subject: facs calibur > > > dear flow cytometrists > > we have been experiencing a problem with our facs calibur recently. the > cytometer cannot > switch from the standby to ready mode. when attempting to acquire data, it > always > comes up withn a warning message " cytometer status not ready". we have > folowed the > trouble shooting guide, and and always ensure taht the machine is pressurised, > and all > connections are intact. what could be the cause of this problem? and how do we > solve it? > > kind regards > danni > >
This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:18 EST