Ray Hicks wrote: "...Calculations in "channel numbers", which need to be transformed into linear units before they're used with any meaning: linear value =10^ (channel value/256)" A term is missing here: the number of decades in the scale. So, the equation should read: linear value =10^ (4*(channel value/256)) Also note that the 256 is the channel resolution of the instrument (or if you are using WinList it is the histogram resolution). So it may be 32, 64, 128, 256 or 1024 depending on your application. This stuff gives us fits too! Albert Donnenberg University of Pittsburgh