Re: Paraformaldehyde and PI

From: Maryalice Stetler-Stevenson (stetler@box-s.nih.gov)
Date: Mon Apr 23 2001 - 16:39:16 EST


It still does not make sense to fix and then us PI for viability.

>Contrary to a variety of posts to this list, while paraformaldehyde
>does indeed
>kill cells, it does not necessarily permeabilize them enough to allow PI into
>all of the cells. For rat bone marrow cells, less than 25% of the
>cells take up
>PI following 0.5% paraformaldehyde fixation (as evaluated under a fluorescent
>microscope), even though 100% of the cells are indeed dead.    Additional
>permeabilization is needed to make all the cells pick up the dye.
>
>There may be some variability between cell lines, etc.  It is ALWAYS best to
>actually look at the cells when a new method is tried.
>
>Carol W. Johnson, DVM PhD
>Pharmacia
>Kalamazoo, MI
>
>
>
>
>
>Maryalice Stetler-Stevenson <stetler@box-s.nih.gov> on 04/18/2001 03:31:37 PM
>
>To: cyto-inbox
>cc:    (bcc: Carol W Johnson/USKZO/PNU)
>Subject:  Re: Paraformaldehyde and PI
>
>
>
>
>PI is only excluded from live cells that are able to maintain their
>cell membranes. Paraformaldehyde does kill cells.Therefore, fixing
>live cells kills them. Then PI accurately demonstrates they are dead.
>Who ever told you otherwise knows absolutely nothing about PI
>exclusion from cells.
>
>
>      Maryalice
>
>
>  >I wonder whether anyone has experience or opinions on this :
>  >
>  >Does 0.5-2 %paraformaldehyde fixation of endothelial cells or tumour
>  >cells permeabilise cells to PI or not.Different sources say yes and
>  >no , but we have found that after removal of adherent cells from
>  >culture plates with 2mM EDTA and fixation that almost all cells
>  >stain with PI . Does paraformaldehyde fixation preclude meaningful
>  >viability assessment or not .
>  >
>  >Many thanks.
>  >
>  ><mailto:mascor@gofree.indigo.ie>mascor@gofree.indigo.ie
>  >
>
>Maryalice Stetler-Stevenson
>Director Flow Cytometry Unit
>Laboratory of Pathology, NCI, NIH
>
>"Learn the rules so you know how to break them properly."
>The Dalai Lama
>
>
>
>PI is only excluded from live cells that are able to maintain their cell
>membranes. Paraformaldehyde does kill cells.Therefore, fixing live cells kills
>them. Then PI accurately demonstrates they are dead. Who ever told
>you otherwise
>knows absolutely nothing about PI exclusion from cells.
>
>
>      Maryalice
>
>
>    I wonder whether anyone has experience or opinions on this :
>
>
>    Does 0.5-2 %paraformaldehyde fixation of endothelial cells or tumour cells
>    permeabilise cells to PI or not.Different sources say yes and no , but we
>    have found that after removal of adherent cells from culture
>plates with 2mM
>    EDTA and fixation that almost all cells stain with PI . Does
>paraformaldehyde
>    fixation preclude meaningful viability assessment or not .
>
>
>    Many thanks.
>
>
>    <mailto:mascor@gofree.indigo.ie>mascor@gofree.indigo.ie
>
>
>
>
>Maryalice Stetler-Stevenson
>Director Flow Cytometry Unit
>Laboratory of Pathology, NCI, NIH
>
>"Learn the rules so you know how to break them properly."
>The Dalai Lama

Maryalice Stetler-Stevenson
Director Flow Cytometry Unit
Laboratory of Pathology, NCI, NIH

"Learn the rules so you know how to break them properly."
The Dalai Lama