Re: sorting onto slides, for morphology...

From: Jan Grawé (jan.grawe@genetics.su.se)
Date: Tue Apr 17 2001 - 04:05:23 EST


Dear Frederic,

using fixed (paraformaldehyde or glutaraldehyde) cells, we sort with
a 500x dilution of our sheath fluid, standard PBS, onto silane-coated
or uncoated slides. This greatly reduces (but does not completely
eliminate) formation of salt crystals. At this dilution, much of the
salt in the drop comes from our isotonic sample buffer, as compared
to the salt in the sheath fluid. This dilution is about the limit for
stable side stream formation on our facstar+. You may have to adust
the dilution for your instrument, and we find that we have to use a
higher drop drive amplitude and twiddle a bit more with amplitude,
stream deflection and phase adjustments than when using undiluted PBS.

After drying, cells are readily visible in the microscope on a slight
background of salt crystals. For photography, we rehydrate the drop
with a few ul of distilled water and add a cover slip, this works
quite well. You may have to mark the position of the sorted drop,
with little salt buildup it can be difficult to locate after drying.

It might also be possible to do this with initially viable cells,
providing they can handle the relatively short period of hypotonic
exposure before the puddle of sorted drops dries.

>will those who sucessfully sort  cells directly onto coated glass slides
>for subsequent morphologic/cytochemical staining /assessment, kindly
>suggest their preferred method to prevent salt crystals from forming?
>kindly also  respond with favorite technique and reagents for fixation.
>
>many thanks
>f preffer
>Frederic I. Preffer
>Department of Pathology
>Charlestown Navy Yard- 7140
>Massachusetts General Hospital East
>Charlestown, MA 02129
>
>voice     [617] 726-7481
>fax        [617] 724-3164

--
Jan Grawé
Inst. för Genetisk och Cellulär Toxikologi
Wallenberglaboratoriet
Stockholms Universitet
S-10691 Stockholm
Sweden

Tel +46-8-163666
mobil+46-70-5667874
fax +46-8-164315
epost jan.grawe@genetics.su.se
http://www.genetics.su.se/gentox



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:14 EST