Re: MEAN CHANNEL VALUE

From: Nancy Perlmutter (ngp@shapirolab.com)
Date: Wed Mar 21 2001 - 10:28:35 EST


Regarding the use of mean channel values on log data:
When data is collected using log amplifiers and displayed on a log scale it
makes sense to use the median rather than the mean channel values as a
method of measuring changes in fluorescence intensity. When data displayed
on a log scale looks like a normal distribution it means that the
underlying linear data is actually skewed toward lower intensity values.
The problem with calculating the mean using log channel values is that it
the mean log channel value is not the same as taking the log of the mean
channel values. The median is a more robust measurement particularly when
working with log channel values. For a much more complete explaination of
the use of mean, CV, median, and robust CV see pages 194-195 of the 3rd
edition of Practical Flow Cytometry by Howard Shapiro. If we are lucky when
Howard returns from his current meeting he will post a much more complete
answer.

Nancy Perlmutter