AnnexinV

From: David_C_McFarland@sbphrd.com
Date: Fri Mar 16 2001 - 10:17:16 EST


I've been following the discussions on AnnexinV and decided to put in my two
cents.  A concept that a lot of people seem to miss is that AnnexinV is more
than useless when used alone, it is misleading.  You have to use it in
conjunction with a dead cell discriminator.  Dead cells stain very brightly for
Annexin V.   And no, AnnexinV+/PI+ double positives do not imply late apoptosis
or apoptotic death.  The PI is only used so that you don't count those cells as
apoptotic.  They are simply dead and that is all you can say about them.  Also,
timing is everything.  AnnexinV is only useful for EARLY apoptosis.  You may
have to do some experimentation to determine the proper time point for
measurement.  If you find you are really interested in late apoptosis, use a
tunel kit  Sometimes you must come to the realization that AnnexinV is not the
best apoptosis assay for your system.  I find, as someone else pointed out, that
adherent cells tend  to give a high percent of background.  Physical
perturbation of the cells can cause PS flipping and false positives. Controls
are paramount.  You need to have a non-apoptotic control that is indeed AnnexinV
negative, or at least dim.  You also need to have a positive apoptotic control
that is indeed Annexin V positive.  If you can't get AnnexinV positivity with a
strong apoptotic stimulus and Annexin V negativity with no apoptotic stimulus,
how can you trust your results?  Your actual test samples may be somewhere in
between and you will have to decide how to interpret them based on your control.
It is not always black and white.  And finally, try to correlate your results
with another assay.  (Caspase 3 is another early apoptosis method.)  I find that
too often experimenters doing flow neglect the microscope.  An easy fluorescence
microscopy technique uses ethidium bromide and acridine orange to distinguish
viable, early apoptotic, late apoptotic and necrotic cells.  (Sorry I don't have
the reference at my fingertips).    I will reiterate what I said before, one
apoptosis assay is not going to work under all experimental conditions.  You may
actually have to try a few things and see what works for you.


And now, just to be Devil's advocate, check these out:

Annexin V Binds to Positively Selected B Cells.  Stacey R. Dillon, et al. J.
Immun.  2001, 166: 58-71.

Annexin V Binds to Viable B Cells and Colocalizes with a Marker of Lipid Rafts
upon B Cell Receptor Activaiton.  Dillon, et al. J. Immun. 2000, 164:
1322-1332.



David McFarland

GlaxoSmithKline