Re: PKH26 for RBC survival study

From: Dennis J. Young (djyoung@UCSD.Edu)
Date: Fri Mar 16 2001 - 14:59:44 EST


PKH26 ex max 551 (& 512), emm 567.

If you have a Innova70C-Spectrum, it's closest line is at 531, but PKH has
a secondary absorption max near the 514 line. The 514 is brighter and may
excite better. How about a 560 or 570 LP for detection?

I would use biotinylated rbc's instead.
 Hoffmann-Fezer G; Trastl C; Beisker W; Berg D; Obermaier J; Kessler W;
       Mysliwietz J; Schumm M; Filser J; Thierfelder S.
     Preclinical evaluation of biotin labeling for red cell survival testing.
   Annals of Hematology, 1997 May, 74(5):231-8.

At 11:42 AM 3/15/01 -0800, you wrote:
>
>I have a researcher who would like to utilize PKH26 as a tracking dye
>for rabbit rbc's survival studies. I have gotten reasonable separation
>of positives and negatives using a "PE" filter setup (555DLP beam
>splitter and 580/30 band pass) with 488 excitation on a MoFlo. The
>researcher is worried about cell toxicity of the PKH26 and wants to use
>as little as possible in his staining procedure. In an attempt to
>maximize fluorescence intensity, I've tried to "tune" the dye head laser
>(INNOVA 70) on the MoFlo to maximum excitation for PKH26 (551 nm) but
>seem to overwhelm the detection PMT with its 580/30 band pass. Any
>suggestions or advice on filter setup or use of PKH26 in this scenario
>is much appreciated.
>
>
---Dennis

Dennis J. Young
Flow Cytometry Core Facility
University of California, San Diego
Internal Medicine Group, Bldg #4, Rm 126
9500 Gilman Drive
La Jolla  CA  92093-0671

Mail:<<mailto:djyoung@ucsd.edu>>
WWW:<<http://cancer.ucsd.edu/SResources/flow.htm>>
Telephone:(858) 822-0407	FAX: (858) 822-0403



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:11 EST