RE: Effect of Lipemia on Side-Scatter

From: Gerhard Nebe-von-Caron (Gerhard.Nebe-von-Caron@Unilever.com)
Date: Fri Mar 16 2001 - 07:56:37 EST


The little bubbles upset your side scatter to give it a huge background signal.
Thus your PMT baseline gets up (constant photon current). Thus if you raise the
voltage you reduce the dynamic range towards saturation of your PMT signal even
further. If you were to look at peak right angle light scatter (RALS) signals
they should all be pretty similar, but your RALS area still show some variation
as it also contains the length of the signal which is size dependent.

Regards
Gerhard




-----Original Message-----
From:	Howard.Gale@med.va.gov [SMTP:Howard.Gale@med.va.gov]
Sent:	Wednesday, March 14, 2001 6:02 PM
To:	Cytometry Mailing List
Subject:	Effect of Lipemia on Side-Scatter


We perform 4-color CD4 panels using BD's lysed, no-wash method on a
FACSCalibur. On heavily lipemic samples the side-scatter signal is
reduced and this lessens the resolution between white cell populations.
Replacing the plasma with FACSFlow sheath fluid corrects this problem.

We can also correct this problem and avoid centrifugation by raising the
side-scatter gains and LOWERING the side-scatter PMT voltage. Raising
the voltage above its usual setting lowers the white cell signals.
This seems paradoxical. It is most noticeable in the granulocyte
population where even the proper shape requires a lower voltage.

Eventually lowering the side-scatter voltage does result in a
decreasing signal just as in non-lipemic samples. Can anyone tell me why
the white cell side-scatter signal has a different response to PMT voltage
changes in the presence of lipemia?