From: Mark Kukuruga (kukuru@med.umich.edu)
Date: Thu Mar 15 2001 - 15:42:45 EST


Juan,
PI has a very broad fluorescence range, and fluoresces well into the 675nm region (see the attached PI ex/em spectrum, borrowed from Molecular Probes).  There's no surprise, then, that you'll see PI signal into "FL3" . . . in fact, you can use this.  By shifting your PI detection to 675nm (FL3), you can a) improve the resolution between PI and FITC, and b) you can then combine PI with PE nicely (usually you'll have to reduce the PI concentration . . .)
This question again illustrates a problem that I've pointed out before concerning a disconnect between a user's awareness of what these FL# designations are, and the actual wavelengths they detect.  I've stated before that we should be aware that these channels are arbitrary, and only defined as specific by the filters we've selected.  So, FL1 is FITC because we use a 530/30 bandpass; FL2 is PE because we choose a 575/25 bandpass, but we can run PI there because "PI has a very broad fluorescence range," FL3 is PerCP . . . 675 nm bandpass . . . and on from there.
It's certainly an advantage having worked with research instrumentation, where any channel is available for any signal.  It gives one a perspective that's often lost when we consider closed-bench systems like the Calibur or the XL.  But, again, a user of flow (especially an operator) should be aware of what these FL# designations are, and that we decided what these channels actually detect by choice of filter.
Juan . . . I'm wouldn't be so concerned that you don't know . . . it is troubling that you say "Nobody in our facility can explain it . . "
MAK.

--
Mark A. KuKuruga, Managing Director
University of Michigan Flow Core
7416 CCGC 0946
(734) 647-3216, fax (734) 936-7376
kukuru@umich.edu


>>> "Juan Oliver M.D." <jao7@columbia.edu> 03/14/01 07:22PM >>>

I have an embryonic pluripotent cell line that has a subpopulation of
cells
with a spontaneous strong signal with a FACScalibur in FL3 (7AAD); the
remarkable finding is that this population of cells increases its signal
one log in
FL3 after PI addition (FL2); 10 micoliters of 0.1 mg/ml into ~ 1ml .
Nobody in our facility can explain it; any suggestions would be
appreciated.  Thanks, Juan Oliver