Richard, Yes, it can. If you look at the Elite, it has essentially the same fluidic and detector configuration as the 753. Using the Sense in Quartz tip (with the focusing lense added) should work. One thing . . . the Elite's detectors may be different (perhaps more sensitive), but I doubt it. With a little work, you'll probably be able to sort as well. MAK. -- Mark A. KuKuruga, Managing Director University of Michigan Flow Core 7416 CCGC 0946 (734) 647-3216, fax (734) 936-7376 kukuru@umich.edu >>> "Salinas, Richard" <rsalinas@saci.org> 03/14/01 04:13PM >>> Greetings fellow flow-ers.I am interested in obtaining some information on possibly using an air-cooled argon laser with our aging Coulter 753. What sort of retro-fitting would need to be done to the flow cell tip? It can't be as simple as removing the current Jet-in-Air tip with a quartz Sort-Sense tip can it?I remember hearing about a company specializing in retro-fitting kits for such a setup, but can't recall their name.Can anyone provide help/suggestions. Has anyone been there/done that?Thanks.Richard A. SalinasCTRC Flow Cytometry Lab San Antonio, TX