Re: Decreased signal in FL3 with SE upgrade

From: Regina Harley (rharley@medicine.umaryland.edu)
Date: Thu Dec 21 2000 - 16:26:10 EST


As he doesn't seem to be the only one with this problem, I'll relay to
you all what I wrote to Andy:

Andy-

Until you mentioned that you had had the PMTs changed, I thought I had a
suggestion. And it may still be something to check out. Apparently BD
uses two different kinds of PMTs. One is more sensitive in the visible
spectrum and is used for the regular fluorochromes, and the other is
more sensitive in the UV range, and is used for that purpose. When we
had our VantageSE installed, the service rep mixed them up and our FITC
PMT was one of the UV kind. Needless to say, the sensitivity of that PMT
was much less than for the PE etc. PMTs. When they switched them back
the proper way, everything was fine. This may be something to look into.

cheers and happy holidays-

Regina

Regina H. Harley, M.S.
Lab Supervisor - Flow Cytometry Facility
Center for Vaccine Development
University of Maryland - Baltimore
685 West Baltimore Street
Baltimore, MD 21201
(410) 706-7376    FAX: (410) 706-6205
rharley@medicine.umaryland.edu


>>> Karim Vermaelen <Karim.Vermaelen@rug.ac.be> 12/21/00 03:40AM >>>

Hi Andy,
We've been experiencing similar problems since we had our Vantage
upgraded to a
Vantage SE : voltages for APC (P6/FL5) and APC-Cy7 (P5/FL4) detection
have to
be set abnormally high (>900) (we use optical bench configuration "6" as
drawn
in the manual). With our old "regular" settings (in the past we
considered 600V
amp to be quite high already), both stained as well as unstained cell
clusters
disappear down to oblivion. I'm sending some dot-plots of this situation
to BD
soon, at this time they don't know any answer to the problem.
FL3 was down shortly after the SE upgrade. However after giving a little
nudge
to the FL1-FL2 vs. FL3 splitter (610 SP) the signal balance was restored
in
favour of FL3 without affecting FL1-FL2 detection.
I assume you have checked your optical bench configuration in detail -
it's
easy to misplace a splitter, invert a filter etc... Refer strictly to
the
configurations drawn in the manual. If this doesn't help it's definitely
up to
BD to fix the problem. I will keep you posted if anything new pops up on
that
front.

"MORSCHAUSER, ANDREW" wrote:

> This message is directed towards FACSVantage users that have upgraded
their
> machine from a FACSVantage to a FACSVantageSE w/Turbosort.  I am
> experiencing a
> decreased sensitivity in FL-3.  When comparing alignment particles
(cRBC,
> Alignflow beads, Calibrite beads) between a FACSVantage, FACSVantageSE
w/TSO
> and
> a FACSCalibur, results have shown the FACSVantage SE signal to be
> significantly
> lower than the Vantage or Calibur.  Furthermore, when looking at
splenocytes
> stained with CD4 PE-Cy7 (787/40 filter), there is no discrimination
between
> negative/unstained cells vs. stained cells.  The signals on the
FACSVantage
> and
> Calibur are nearing the third decade while the FACSVantageSE signal
was
> negative
> with log amps and voltages greater than 900V.  I tried increasing the
log
> offset
> as far as it will go and found little if any improvement in the
signal.
>
> Boards have been replaced and new PMTs in FL3 have done nothing to
remedy
> the
> problem.  BD has been very responsive to my concerns, but the problem
is
> still
> present and I was wondering if anyone else has experienced similar
problems
> with
> the extra mirror and different optical path for a 5-color FACSVantage
SE.
> Any
> comments or suggestions would be greatly appreciated.
>
> Andy Morschauser