Re: Dead cell exclusion in FL4

From: Derek Davies (daviesd2@icrf.icnet.uk)
Date: Fri Oct 13 2000 - 05:25:25 EST


Hi Karim,

I use TO-PRO-3 extensively here as a dead cell discriminator. We have a
couple of dual laser Caliburs and being able to get away from always
using PI or 7AAD has freed up the other fluorescence channels for other
markers. As for sensitivity, I did a lot of checking that TP3 labelled
the same cells as PI and it is assumed that it works in the same way ie
it will only enter cells with compromised membranes and will
intercalate into the DNA helix. One thing that I did find though was
that on our sorters where we use higher laser powers at 488nm, we are
seeing some excitation of the TP3 there and emission from the primary
beam. On the Calibur, the signal is slightly less intense than PI but
this isnt really an issue. I would normally use a stock of about 1uM and
add 10ul per ml.

It is not normally necessary to RNase treat cells before using any
DNA-binding dye as a dead cell discriminator as the dye should be
excluded from the live cells.

Hope that helps,
Derek


On Thu, 12 Oct 2000, Karim Vermaelen wrote:
> 1. We'd like to do some 4-5 color analysis with dead cell exclusion in
> FL4. Anybody has any experience with TOPRO-3 and sorts (Mol. Probes)?
> How does it compare (sensitivity/specificity) with PI or 7-AAD?
> 2. When doing dead cell exclusion, is it necessary to RNAse-treat cells
> before applying the DNA-probe (like you would do for cell cycle
> analysis); we're working with highly activated cells with high RNA
> synthesis and we're concerned about possible uptake of the
> nucleotide-probe by mRNA.
>
> Thanks a lot!
>
> Karim Vermaelen

************************************************************************
Derek Davies                       Voice: (44) 020 7269 3394
FACS Laboratory,                   FAX: (44) 020 7269 3100
Imperial Cancer Research Fund,     e_mail: derek.davies@icrf.icnet.uk
London, UK			   mobile: 07790 604112

Web Page: http://www.icnet.uk/axp/facs/davies/index.html

In tenebris lux
*************************************************************************



This archive was generated by hypermail 2b29 : Sat Mar 10 2001 - 19:31:34 EST