Re: Cryopreservation of bags and vials

From: Scott Burger (burge009@tc.umn.edu)
Date: Tue Oct 03 2000 - 18:03:38 EST


At 08:17 PM 10/02/2000 -0400, Liza McGuire wrote:
>Is there any possibility to freeze bags and vials in a rate control freezer
>simultaneously? There are many issues why this would not work, but I
>wonder if anyone ever tried it and can provide any input.

Our standard method for cryopreserving all our clinical products involves
freezing bags and vials together in a programmable freezer.  The product
bag is held in a metal press, with the temperature probe placed next to the
bag.  A rack containing sample vials from the product rests on the metal
press, and the vials and product bag are frozen together.  Cell recovery
from the cryovials is not quite as high as from the product bag, but still
very acceptable.

You may want to look at Jershin, et al., Hematopoietic progenitor cells
frozen in ampules are satisfactory reference vials for bulk freezing.
Progress in Clinical & Biological Research. 389:657-61, 1994.

Best of luck,

Scott

--------------------------------------------------
Scott R. Burger, M.D.
Director, Cell Therapy Clinical Laboratory
University of Minnesota, Fairview-University Medical Center
420 Delaware Street S.E., Mayo Mail Code 609
Minneapolis, MN 55455
612-626-4919
612-625-2199 - Fax
burge009@tc.umn.edu
http://www.mbbnet.umn.edu/labs/ctcl.html



This archive was generated by hypermail 2b29 : Sat Mar 10 2001 - 19:31:33 EST