From: Howard Shapiro (hms@shapirolab.com)
Date: Thu Jul 27 2000 - 18:01:45 EST


Andrew Beernink wrote-

>Can anybody fill me in on the use of 543 nm HeNe lasers for flow?
>Specifically, is 1-3 mW enough power for most applications, or does
>someone make one in the 10 to 50 mW range?  Aside from PE and its
>conjugates, what will this excite?

As far as I know, 3 mW is about as much as can be gotten out of a 543 nm
HeNe laser.  For PE excitation, this is at least as good as 15 mW at 488
nm, and there is much less excitation of autofluorescence from mammalian
cells at 543 nm than at 488 nm, so the stained cells and the unstained ones
are further apart.  The green He-Ne works well in instruments with very
efficient light collection; Beckman Coulter offers one as an option in the
Elite and Altra cell sorters, and, while B-D has not offered the option in
the FACSCAlibur, the FACSCount, a relatively inexpensive dedicated flow
cytometer for CD4/CD8 counting, does use a green He-Ne.  The green He-Ne
laser would be less useful in a typical stream-in-air sorter, such as a B-D
FACSVantage or Cytomation MoFlo, because their light collection is somewhat
less efficient.  Green frequency-doubled YAG lasers, emitting at 532 nm,
would be a good choice here and in other instances where higher power is
needed; they are small and energy efficient, and units with up to several
watts of power output plug into a standard 110V outlet, draw under 100
watts, and are air-cooled.  In fact, the availability of Green YAG lasers
is probably keeping manufacturers from trying to build more powerful green
He-Ne's, which would probably be at least as expensive and substantially
larger.  Also in their favor, single-mode green YAG lasers have very low
noise, and are better for scatter measurements than the noisier green He-Ne
lasers.


Both 532 and 543 nm are good for excitation of PE and its tandems, also for
Cy3, rhodamine dyes, pyronin Y, some of the Molecular Probes POPO, BOPRO,
etc. cyanine nucleic acid dyes and Alexa protein labels, ethidium,
propidium, 7-aminoactinomycin D, LDS751, the DiICn(3) cyanine dyes
(membrane potential sensitive for n  6 or less, tracking dyes for n 14 or
more), PKH26, and resorufin-based fluorogenic substrates.  For many of
these, 532 if preferable to 543 because the Stokes shifts are relatively
small, as well as because more excitation power is available from Green YAG
lasers than from green He-Ne lasers.

On the negative side, neither 532 nor 543 nm lasers will excite fluorescein
or its derivatives, GFP and its genetically modified variants, or acridine
orange, and both wavelengths would interfere (532 more than 543) with
measurement of these materials using a second 488 nm beam, although some
games might be played with emission filters to minimize the interference.

-Howard