Re: BrdU/PI cell cycle analysis

From: Mark A. KuKuruga (kukuru@umich.edu)
Date: Fri Apr 28 2000 - 07:59:16 EST


Diane,
Some are of the opinion that log amplification of the BrdU labeling may be too sensitive,
thus revealing some low level, and possibly non-chromatin DNA, content.  We've looked at
this both ways, and generally find the numbers to correlate.  It is true that with a
direct labeling technique the signals are considerably lower than with an indirect
system,
so linear display of the BrdU is workable.  If, however, your labeling technique gives a
very bright result, you may be forced to use a log display to manage it.
Again, it's probably a good idea to look both ways, and see for yourself if there are
significant differences.
MAK.

Diane M Sharp wrote:

> Dear Flow Group,
>
> I have seen several publications lately that show BrdU cell cycle analysis, but show
> both PI and FITC channels on a linear scale.  I generally look at FITC on a log scale.
> Does this affect results in any way?
> Thanks for any comments.
>
> Diane Sharp
> DuPont Pharmaceuticals Company
> Diane.M.Sharp@dupontpharma.com

---
Mark A. KuKuruga, Managing Director
University of Michigan Core Flow Cytometry
<http://www.cancer.med.umich.edu/flow_cytometry>
phone: 734-647-3216  fax: 734-936-7376
kukuru@umich.edu



This archive was generated by hypermail 2b29 : Sat Mar 10 2001 - 19:31:18 EST