RE: Sorting CD4+CD3+ cells

From: Newsom, Brian S. (BSNEWSOM@txccc.org)
Date: Fri Apr 14 2000 - 18:40:12 EST


Joy,

 For sorting purposes you will not be able to use the CD3/8/4 set from BD
unless the CD4 is a PerCP-Cy5.5 conjugate, the regular CD4 PerCP does not
show up on the sorter due to the high laser power (unless you are using a
FACSort). The Coulter reagents should work if you have the correct filters.
Some filter sets do not work well in differentiating the PE/ECD/PC5
combination-and the CD45 is probably unneccesary.

Here is what you need to do in a nutshell: gate on a CD3 vs. SSC to get a
popluation of CD3+SSC(low) cells (T-cells), you can also gate on FSC vs SSC
in addition to clean it up a little more. Using that gate look at CD4vsCD8
and you should be able to pick out the populations of choice. Your logical
gating strategy for the sorter will be the same as it is on an analyzer.
Gate on CD3+/SSC(low)/CD4+ cells to sort one way and CD3+/SSC(low)/CD8+
cells to sort the other way. Unless you have a MoFlo 4-way sorter you will
have to do two separate sorts since you want three populations. The third
population will be a gating strategy to isolate CD3+/SSC(low)/CD4-CD8-
cells.

For fluorochromes, I would suggest  FITC/PE/PE-CY5 or FITC/PE/PerCP-Cy5.5.

Brian Newsom
Technical Application Specialist
BD/Pharmingen

-----Original Message-----
From: Joy Mundy
To: cyto-inbox
Sent: 4/13/00 7:45 PM
Subject: Sorting CD4+CD3+ cells


Hi everybody,

I am passing the following message on behalf of a researcher in our HIV
research labs. In our lab we are familiar with the analysis of T cell
subsets, currently using the Coulter tetraCHROME reagent. However this
researcher is interested in cell sorting and how to set up the BD flow
sorter to get the cells of interest. As a consequence I am of no use
whatsoever. The antibodies that we currently have at our disposal are
the
Coulter tetraCHROME CD45 FITC/CD4 PE/CD8 ECD/CD3 PC5 and BD CD3/CD8/CD4
together with the BD Tritest control. Analysing the cells are no problem
it's just how th set up the sorter to get the correct cells out!

This is the problem;

Basically what I want to do is as follows
Take 40mls patient blood, ficoll separation.
Stain PBMCs in such a way that I can put them thru a cell sorter and
collect as many cells as possible as need to extract DNA from the
following , to then do a series of PCR reactions to determine integrated
HIV DNA.
Interested in CD4+ lymphocytes (CD3+/CD4+), CD8+ and the small number of
double negative cells (cd4-/CD8-) as these are possibly CD4+ cells that
have down regulated their receptor post HIV infection. I guess the ideal
scenario is to pass the cells thru and collect all CD3+ cells and then
the rest. Then pass thru these CD3+ cells and differentiate between the
4s and the 8s????

Thanks any suggestions welcome !

Joy Mundy
Division of Human Immunology
I.M.V.S.
PO Box 14
Rundle Mall PO
Adelaide 5000
Australia