Re: FACSComp QC

From: David L. Haviland (dhavilan@imm2.imm.uth.tmc.edu)
Date: Thu Mar 30 2000 - 13:08:34 EST


Tami Rosario wrote:
>
> I have a question for those of you who run FACSComp on the FACSCaliber
> for QC.  Do you run this each time you start up the cytometer or only
> occasionally.  I have been told that I should run FACSComp each day that
> I run samples, but the beads are fairly expensive.  Any alternatives?

Tami:

Glutaraldehyde Fixed Chicken erythrocytes, get them from Sigma.  They're
cheap and one vial will last nearly forever.  Given the use of our
Calibur, I check about once every week and a half, or unless a problem
is reported (clog).  When I've kept them at room temp, on the bench, I
find that the cv's and fluorescence intensity has actually improved
after about a week.

My great BD service rep simply keeps them at ambient temperature in his
tool kit.

Hope this helps,
David
===========