Re: PMT for Cy7?

From: Marty Bigos (bigos@stanford.edu)
Date: Thu Mar 30 2000 - 10:55:12 EST


Hi Joseph-

The R3896 from Hamamatsu is a red-sensitive PMT that is
plug-compatable with the Facstar+. You should be able to order direct
from them. Here it costs US$650 each in small quantities.

Before you do this, you could check and see what PMT is now in FL5.
Maybe BD already gave you one (ha, ha, only kidding).

Seriously you should check and see what kind of PMTs you do have.
Although the 3896 is "red sensitive", it is also more sensitive
throughout the spectrum at lest up through green. Upgrading your PMTs
is one of the cheaper ways to improve your measurement sensitivity.
Your FL3 (if you use it for PE-Cy5 or PE-TR) and FL4 (APC) will
probably noticeably benefit, especially if they are R928 PMTs.

Also, I am confused about what you are using for filters. I am
assuming your 710LP is the beam splitter between the two channels.
You will also need a bandpass filter in the APC channel to block any
scattered laser light.

Marty Bigos
Stanford Shared FACS Facility

>We want to try out APC/Cy7 on our FACStar Plus, some clues
>would be helpful from anyone who has used it on a StarPlus
>or Vantage.
>
>We have a red HeNe laser that gives 35-40mW.
>We have ordered Omega filters 710DRLP and 740EFLP to read
>APC and APC/Cy7 in FL4 & FL5.
>
>I believe the standard PMT will probably not be red-sensitive
>enough for Cy7.
>If it won't work, what PMT should I buy?
>Also what's the cheapest source for us in Australia?
>
>What have I missed or forgotten?
>
>Thanks, Joseph.
>
>Joseph Webster, Flow Cytometry Facility,
>Centenary Institute
>Locked Bag No.6, Newtown, NSW 2042, AUSTRALIA.
>Ph: 61-2-9565-6110	Fax: 61-2-9565-6101
>e-mail: J.Webster@centenary.usyd.edu.au



This archive was generated by hypermail 2b29 : Sat Mar 10 2001 - 19:31:14 EST