Re: viability on fixed cells

From: Julie Oughton (oughtonj@ucs.orst.edu)
Date: Thu Mar 16 2000 - 13:46:55 EST


Yes--it is possible! I had excellent results using 7-AAD to stain dead
cells prior to fixation.
The leakage of 7-AAD to live cells can be blocked by using non-fluorescent
actinomycin D in the media. I followed the technique outlined by Schmid et
al. (1999) in Cytometry 35:64-74.

Julie Oughton
Oregon State University


>
>
>