Re:FITC-VAD-FMK marker for apoptosis

From: DARZYNKIEWICZ ZBIGNIEW (DARZYNK@nymc.edu)
Date: Wed Mar 08 2000 - 17:33:21 EST


To the ongoing discussion on the use of fluorochrome-labeled protease
inhibitors to detect activated caspases let me add a note about our
experience.
First of all, the use of labeled inhibitors that bind to active centers of
proteases has long history. In the 60's this approach was applied to detect
active proteases e.g. in mast cells with radioisotope-tagged protease
inhibitors which were then traced by autoradiography (see Nature, 213,
1198-1203, 1967). The essence of the methodology is based on covalent
binding of the inhibitor (usually at 1:1 stoichiometry) to the active center
of the enzyme. In the case of caspases one of several ligands that may
specifically bind to their active center is a fluorochrome labeled
fluoromethyl ketone-peptide conjugate. Such ligands [e.g.
benzyloxycaronylvalylalanylaspartic acid fluoromethyl ketone (FAM-VAD-FMK),
or FITC-VAD-FMK] have recently become comercially available. The first is
available from Promega, Madison, WI, the second from Intergen Co., Purchase,
N.Y. These ketone reagents penetrate through the plasma membrane of live
cells and are  relatively nontoxic to the cell. Their irreversible binding
to active centers of the caspases ensures that only the cells with the
activated enzymase become labeled. Because these inhibitors consist of three
amino acid recognition sequence instead of four (four amino acid recognition
is typical for individual caspases) they are rather generic for most
caspases than specific for the particular one. We have experience with the
reagent offered by Intergen. This reagent performs very well in two model
systems of apoptosis, one of HL-60 cells treated with camptothecin
(apoptosis induced via mitochondrial pathway) and another, with the same
cells treated with TNF alpha plus cycloheximide (death ligand iduced
apoptosis). We did measure cell fluorescence by laser scanning cytometry and
confirmed, by cell re-location and morphologic examination, that that the
cells labeled with FAM-VAD-FMK were apoptotic (manuscript in preparation).
We also have a good experience with the antibody that detects the product of
PARP cleavage that is offered by Promega (Exp. Cell Res., 255, 125-132,
2000).

Zbigniew Darzynkiewicz
	]



This archive was generated by hypermail 2b29 : Sat Mar 10 2001 - 19:31:11 EST