Re: sorting bacteria

From: Howard Shapiro (hms@shapirolab.com)
Date: Thu Mar 02 2000 - 19:20:23 EST


Jackie Saleh writes:

>...could someone recommend alignment beads for
>bacterial sorting. My
>instrument is a FacsVantage.

In my experience, if you can trigger successfully on forward scatter
signals from 0.5 um polystyrene beads, you can trigger successfully on
forward scatter signals from bacteria such as Staphylococcus aureus,
Escherichia coli, or anything bigger.  This may require some adjustment of
your scatter optics.  I am assuming you will be using a water-cooled laser;
scatter sensitivity tends to be better because there is usually less
optical noise on a water-cooled than on an air-cooled laser (which has
nothing to do with the cooling; it reflects a difference in power supply
design).

If you are using 488 nm excitation and measuring green fluorescence (which
you would do for GFP, and I'm guessing that a high-tech pharmaceutical
company would be at least as interested in sorting bacteria on the basis of
GFP fluorescence as on any other basis), you can use 0.5 um Polysciences
yellow-green calibration beads; they have very good (2% or lower) CV's in
both forward scatter and fluorescence. These beads also come in smaller
sizes (near 0.25 um); if you can trigger on those, you're really in good shape.

-Howard