Fluorescence Quantitation by Flow Cytometry.


Ricardo Morilla
Academic Department of Hematology and Cytogenetics.
Royal Marsden Hospital.
Fulham Road London SW3 6JJ.
e-mail: ricardo@icr.ac.uk

View this document as a Microsoft Word file.
View this document as an Adobe Acrobat file.

Definition and introduction:

Fluorescence quantification by Flow Cytometry is defined as the measurement of the intensity of staining of cells and provides an absolute value for the light intensity it measures.

Quantification of fluorescence is performed comparing cell fluorescence with a known external standard. By using different beads commercially available, it is now possible to measure the quantity of fluorescence relative to the peak channel obtained by flow cytometry, using a standard curve for its calculation.

There are two units to express the fluorescence quantification, the ABC (Antigen Biding Capacity) and the MESF (Molecules Equivalent Soluble Fluorochrome).

Applications of quantitation:

The quantitation of fluorescence molecules by flow cytometry provides additional information which is useful for the precise characterization of cells.

It is useful in hematology for the identification of hemopoietic cell populations, both normal and leukemic. In certain hematological malignancies, precursor cells associated antigens can be under or over expressed on the malignant cells and can therefore be regarded as a leukemia-associated feature when compared with the normal counterparts. [6]

There is also a hierarchy of expression that is characteristic of different lineage, i.e.: dim expression of CD4 in monocytes and brighter expression in lymphocytes.

Some antigens show unimodal expression on blood cell populations, other antigens may be heterogeneously distributed with different densities. This may indicate functionally different subsets in relation to a particular differentiation stage of the cells.

The values of antigen density, when added to percentages or absolute counts of positive cells, exploit the informative value of the immunofluorescence test in the interest of defining both normal and abnormal differentiation pathways and subset compositions as well as signs of cellular activation. [4]

Quantitative flow cytometry has been used in the study of maturation processes looking at differentiation antigens, in activation of neutrophils and complement receptors, in functional assays of adhesion molecules, in infections looking at the expression of virus receptors, in oncogene products, drug receptors, steroids receptors, immunophenotyping and pathological situations.

Procedure:

Type of quantification beads' kit
Preparation of specimen and beads
Data acquisition
Calculation of ABC or MESF values

Type of quantification beads' kit:

The choice of the type of beads depends on the procedure used for the sample preparation.

The general principle of fluorescence quantification by beads is the same for different methods of staining of cells.

This is based in commercially available kits. In these kits, there are usually two tubes. One tube with a mixture of four beads, with four different levels of uptake of fluorescence, one very dim, one very bright and two intermediates, and another tube with a blank, i.e. beads with no uptake of fluorescence.

Direct immunofluorescence:

QSC (Quantum Simply Cellular): These are beads coated with goat anti mouse immunoglobulins, each level of the standard can bind to a certain amount of mouse Ig. They are used to quantify direct immunofluorescence. They measure ABC (Antibody Binding Capacity), when saturated with the same fluorochrome conjugated monoclonal antibody as used on the cell sample.

These beads require one calibration for each monoclonal antibody.

Quantibrite: These are beads manufactured only by Becton & Dickinson to be used with their Cell Quest program Quanti Quest. These are beads labeled only with PE. There are not many McAb available at the moment. BD claims that they can label the PE at a ratio 1 to 1.

Indirect Immunofluorescence:

QIFIKIT these are beads coated with monoclonal antibodies which mimic monoclonal antibody bearing cells. These receptors bind to the secondary antibody used both to stain the cells and the beads, therefore they require only one calibration per experiment.

Direct and Indirect immunofluorescence:

FCSC (Quantum beads): These beads are coated with known molecules of fluorochrome, they are available conjugated to FITC or PE and measure the MESF. The fluorescence of the cells is compared to standard molecules of fluorochrome.

Preparation of specimen and beads:

The samples should be prepared as usual by the method of choice of the laboratory.

Please note that if one is doing intracytoplasmic staining, the beads should not undergo the process of fixation and permeabilisation.

i. Direct immunofluorescence:

The beads are incubated with the relevant McAb to be quantified for one hour at 4oC. Kits used for this methodology are: Quantum Simply Cellular, Quantibrite and FCSC Quantum Beads.

The specimen is prepared following the standard in house protocols.

One standard curve is required for each McAb to be tested.

ii. Indirect immunofluorescence:

The beads are stained at the stage of applying the secondary antibody, they are incubated for one hour at 4oC. The cells of the specimen, are stained following the in house protocols.

One standard curve is sufficient for different McAb tested, provided that the same secondary antibody was used.

The monoclonal antibody is not directly conjugated to a fluorochrome and a second incubation with a fluorescent antibody against the primary is needed to detect the reaction by flow cytometry. The secondary antibody can be labeled with one fluorochrome, i.e.: FITC, PE or Third color. [1]

Fig. 1: Histogram showing blank and 4 fluorescence peaks. Statistics showing peak channel values

Quantification rules:

There are three essential rules to perform quantification:

a. The monoclonal antibody has to be applied at saturating amounts both for the beads and the cells in the specimen.

b. The same reagent , from the same company and at the same dilution should be used throughout the experiment and other future analysis.

c. The instrument fluorescence setting should be maintained unchanged once the beads have been run and the analysis of the unknown sample should be acquired with the same setting.

Data acquisition:

Acquiring beads:

The tube with the beads is acquired firstly in all cases where one experiment is sufficient for the quantification, this is the case for the Quantum beads and the QUIFIKIT beads. In the case of the Quantum Simply Cellular, where a tube is run for each monoclonal antibody, the tube with the beads for that particular McAb should be run first, and then all the other tubes with beads for the other different McAb.

The SSC voltage needs to be decreased more than that for cells in order to bring the beads into the FSC/SSC dot plot. There could be some doublets if the tube was not shaken vigorously, these are excluded by doing a tight gate around the beads, data is acquired in these gated beads.

The instrument should be set up in a way that the fluorescence signal of the tube with the blank (unlabelled) beads is located in the region between 0 and 101, and four other peaks of fluorescence should be seen along the axis of the relevant fluorochrome. Once the fluorescence voltage of the instrument is set up, these settings are maintained throughout the rest of the analysis of the unknown samples. In the case of the QSC, the settings for each McAb should be used accordingly.

Acquiring the sample:

The SSC voltage will need to be icreased to obtain the cells profile within the FSC/SSC dot plot.

The settings for the fluorescence channels used for the beads should be maintained throughout the rest of the analysis of the tubes of the sample.

Calculation of ABC or MESF values:

Relevant software is provided with the quantification kits. These programs are user friendly. The data obtained from the cytometer is put into the program and automatically a standard curve is produced. These programs take into consideration the make of the instrument used, the voltage for that sample, the fluorochrome used, the supplier of the McAb, etc.

The program calculates the ABC and/or MESF value of the unknown sample and saves the data.

The standard calibration curve is obtained by plotting the values of the peak channels of the blank and the other four peaks obtained from the flow cytometer against the assigned known number of molecules of fluorochrome obtained from the supplier of the beads.

The peak value of the unknown sample is obtained by running the sample with the same fluorescence setting as the beads minus the peak value of the control tube (tube with the isotypic conjugate fluorochrome but with no primary antibody)

Using the software provided with the beads, fill in the relevant data and enter the peak channels of the different peaks for the beads obtained from their analysis on a histogram plot to obtain a standard curve.

Pitfalls of the method:

One should have careful consideration to the fact that the volume of McAb needed to saturate the cells sample, is not necessarily the same as required to saturate the beads and a greater volume of McAb may be needed to saturate the latter.

The position of the fluorescence setting of the blank may have to be negotiated and set up slightly lower or higher than the ideal situation just within the 101 limit. Some samples have very dim fluorescence and if the blank is set up too low, the intensity of the sample may be difficult to calibrate.

Overestimation of cell bound ABC values can occur when QSC beads are used because epitopes on cells are homogenous and of high affinity while bead bound GAM Ig bind mouse Ig with variable affinity.

Quantification clinical applications in hematology:

Acute Lymphocytic Leukemia:

The leukemic cells represent the malignant counterparts of normal hematopoietic precursors expressing terminal deoxynucleotidyl transferase (TdT), CD10 and CD19.

Normal TdT+ precursors have significantly higher number of TdT (>100x 103) and lower number of CD10(<50x103) and CD19 (<10x103).than in B lineage ALL blasts TdT (<100x103) CD10 >50x103) and CD19 (>10x103) molecules per cell. These differences were statistically highly significant, therefore, the quantitative analysis of TdT combined with CD10 and CD19 may allow a clear distinction between normal precursors and minimal residual leukemia in B-lineage ALL and avoid the pitfall of misinterpreting regenerating B-cells as evidence of relapse.[2]

Quantification of B-cell antigens:

The B-lymphocytes of B cell chronic disorders have their counter part in the normal B-cells of the peripheral blood. Several published works studied a series of B-cell antigens in to assess if there was any difference between the normal B-cell in the blood and those B-cells from different chronic B-cell disorders. Also to assess if there was any distinct expression pattern in antigen binding capacity amongst different B-cell disorders to help in diagnosis, detection of minimal residual disease, and sometimes treatment. [7]

Quantification of T-cell antigens:

Most T-cell antigens are expressed on normal and neoplastic T lymphocytes and for this reason it is not easy to distinguish between the immunophenotype of normal and malignant T-cells.

There are results published that show that the quantitative analysis of CD3 and CD7 and their combined evaluation may enable a distinction between normal and leukemic T-cells and could facilitate the monitoring of minimal residual disease.

The study has also defined the T prolymphocyte as a cell of intermediate maturity between thymus derived and peripheral T lymphocytes.[3]

Conclusions:

Quantification is a useful methodology to extract additional information from the flow cytometer.

This extra information is helpful in hematological malignancy and provides a refined knowledge of hemopoietic differentiation, permits a more objective definition of positivity and finally it helps to ascertain malignancy, and normality.

It has also been shown that it provides an additional parameter for the detection of MRD, particularly in B-cell ALL where the outcome of treatment could be predicted.[5]

References:

1. Poncelet, P., George, F., Papa, S., and Lanza, F.
Quantitation of hemopoietic cell antigens in flow cytometry.
European Journal of Histochemistry. 40/Suppl. 1. 15-32 (1996)

2. Farahat, N., Lens, D., Zomas, A., Morilla, R., Matutes, E., and Catovsky, D.
Quantitative flow cytometry can distinguish between normal and leukemic B-cell precursors.
British Journal of Hematology 91, 640-646 (1995)

3. Ginaldi, L., Matutes, E., Farahat, N., De Martinis, M., Morilla, R., and Catovsky, D.
Differential expression of CD3 and CD7 in T-cell malignancies: a quantitative study by flow cytometry. British Journal of Hematology 93, 921-927 (1995)

4. Bikoue, A., George, F., Poncelet, P., Mutin, M., Jannosy, G., and Sampol, J.
Quantitative analysis of leukocyte membrane antigen expression: Normal Adult values.
Cytometry ( communications in Clinical Cytometry) 26: 137-147 (1996)

5. Farahat, N., Morilla, A., Owusu-Ankomah, K., Morilla, R., Ross Pinkerton, C., Treleaven, J., Matutes, E., Powles, R., and Catovsky, D.
Detection of minimal residual disease in B-lineage acute lymphoblastic leukemia
by quantitative Flow Cytometry.
British Journal of Hematology 101: 158 164 (1998)

6. Ginaldi, L., De Martinis, M., Matutes, E., Farahat, N., Morilla, R., and Catovsky, D.
Levels of expression of CD19 and CD20 in Chronic B-Cell leukemias.
Journal of Clinical Pathology. Vol 51: 364-369 (1998)

7. D'Arena, G., Musto, P., Cascavilla, N., Dell'Olio, M., Di Renzo, N., and Carotenuto, M.
Quantitative flow cytometry for the differential diagnosis of leukemic B-Cell chronic lymphoproliferative disorders.
American Journal of Hematology 64:275-281 (2000)