Re: meaning of FL2-area and FL-2 width

From: Larry Arnold (lwarma@med.unc.edu)
Date: Tue Jan 08 2002 - 16:48:09 EST


Janet

Parameter height, area, and width can be visualized by looking at the curve
(pulse) one sees if looking at (or envision) an oscilloscope tracing of a
particle transiting the laser beam.  The height is the height of the pulse,
the area is the area under the curve and the width is the width of the
pulse at some threshold level (for linear signals only).  The area always
provides the total fluorescence of the cell so for DNA measurement this is
what we want.  If the laser beam is wider than the cell the height and area
are proportional.  If the laser beam width is less than the cell diameter
(slit scanning) then this is not true and you need to use area to get the
total fluorescence.  Envision 2 identical cells with the same number of
fluorescent molecules attached.  In one cell the fluorochrome is
concentrated in one small area whereas with the other cell the fluorochrome
is evenly distributed over the entire cell.  In the first cell (using a
narrow laser beam and disregarding any fluorochrome interactions which may
quench some fluorescence) the height of the fluorescence signal will be
very large but the width will be very narrow.  In the second cell the
height will be much less but the width will be greater.  Using height one
would conclude that the first cell has more fluorescence which of course is
untrue as we set up the experiment.  Using the area both cells will be
evaluated as having the same fluorescence.  Now back to your situation
where the laser beam is almost certainly wider than the cell (nucleus)
diameter.  Here width for a single cell is actually the width of the laser
beam.   Envision a single G0 cell transiting the laser compared to a G0
doublet (with the doublet longest axis perpendicular to the laser
beam).  The height of the doublet (in a FACScan) will be slightly greater
than that of a singlet (if the laser beam diameter was the size of a
nucleus then the height would be exactly the same since no more than one
nucleus worth of PI could be in the laser beam at any one time).  Note that
these conditions change as the size of the cell (nucleus in the case of
DNA) changes since the laser beam diameter is fixed.  However the width of
the doublet compared to the singlet will be much greater (i.e. up to
twice).  The area of the G0 doublet will also be twice that of the
singlet.  Without any doublet detection/correction you will measure this G0
doublet as a single G2/M cell - a significant error.  By plotting the width
versus the area you can visualize the G0 doublets (and higher order
complexes) from the singlet G2/M cells and gate them out.  (You can also
use height vs area but width vs area on the FACScan/Calibur seems more
sensitive).  This is hardware doublet discrimination.  Software methods
(e.g. ModFit) calculate (and then subtract) the number of G0 doublets by
counting the triplets, quadruplets etc.  However, this often can not be
used depending on the cell populations present (for example with
endoreduplicating cells).  Note that the doublet G0 population width is a
smear (on a histogram) rather than a tight population.  This is because not
all doublets will transit the laser beam with their longest axis
perpendicular to the laser. As the angle between the longest axis and the
laser beam becomes less than 90 degrees then the width decreases.  If the
doublet long axis is parallel (or close to parallel) to the laser beam the
doublet is not detected.  Thus, the fluidics system needs to be operating
properly - i.e on the FACScan running on the low fluidic setting (this is
critical for the best CV also as well as separating independent doublets (2
cells not connected but in the laser at the same time)).

Hope this helps.

Larry


At 01:41 PM 1/7/2002, you wrote:

>I know this is going to sound like a stupid questions but we(me and several
>post docs) are currently involved in a heated debate as to the nature and
>meaning of FL2-area and FL2-width in connection with the measurement of
>DNA content(PI stained) of ethanol fixed cat lymphocytes.  Can someone give
>us a good definition of each and an explanation of why we use these for DNA
>measurement instead of FL2-Height?  We are using a BD FACSCalibur for our
>work.  I have done DNA work for years and I have never really thought about
>this before-this is how I was taught to do it.  Again I am sure this is
>result of never having taken a formal course on flow and just picking it up
>from the person who previously held my job and other investigators.
>
>I have searched the archive and can not find a good explanation.  Maybe I
>need a course in using the search engine for the archive!!!
>
>
>Thanks in advance.
>
>Janet Dow
>
>Janet Dow
>Research Technician and Manager
>Flow Cytometry Facility
>North Carolina State College of Veterinary Medicine
>Room C-314
>Raleigh, NC 27606
>(919)513-6364

Larry W. Arnold, Ph.D.
Associate Professor
Director, Flow Cytometry Facility
Department of Microbiology and Immunology
Lineberger Comprehensive Cancer Center
CB# 7290
University of North Carolina at Chapel Hill
Chapel Hill, NC 27599
Phone: 919-966-1530
FAX: 919-962-8103



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:59:19 EST