Compensation question

From: Andrew Oberyszyn (oberyszyn.2@osu.edu)
Date: Wed Jan 02 2002 - 12:06:36 EST


Hi Flow people!
First let me wish everybody Happy Holidays and keep up the great work on
this newsgroup!

I ran some 3 color samples (CD3 FITC/CD4 PE/CD8 PE-Cy5) on my Beckman
Coulter Elite today and I saw something that didn't make sense to me.  I've
attached three slides of the histograms that I got.

I ran the single color positives and compensated so that the mean values
where correct (see Slide 1, compensation settings #1).  I then ran the
samples (example of one on slide 2).  There where 2 populations that didn't
seem right to me (looked like they where undercompensated).  I increased
the compensation so that it "looked right".  I then ran the single color
controls again using the increased compensation values (slide 3).  The mean
values I got seemed to be overcompensated.

I apologize for being somewhat ambiguous in the descriptions above (I'm
trying to at least get out of here on time today!) but the histograms in
the attachment are more informative.

I've done three color work before but have never come across this
problem.  The user says that the antibodies are the same (i.e. not
different lots within an antibody throughout the samples).

I realize that alot of people are on vacation already and responses to this
might be sparse but any help would be greatly appreciated (Mario?  Help!)

Thanks in advance!
Andy

(:)-(:)-(:)-(:)-(:)-(:)-(:)-(:)-(:)-(:)-(:)-(:)-(:)-(:)
   Andy Oberyszyn, M.S.
   The Ohio State University
   Analytical Cytometry Laboratory
   416 Comprehensive Cancer Center
   410 West 12th Avenue
   Columbus, Ohio 43210
   Tel: 614/292-FLOW(3569)
   Fax: 614/292-7335
   E-Mail: cytometry@osu.edu

(-):(-):(-):(-):(-):(-):(-):(-):(-):(-):(-):(-):(-):(-)

"What if the Hokey Pokey is really what it's all about?!?"

Slide1.JPG
Slide2.JPG
Slide3.JPG



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:59:19 EST