Braun, I can vouch for this. I had a similar problem a few years ago and Dr. Haugland pointed this phenomenon out to me. I was looking at small bacteria binding to large cells and the shift in the flow cytometer's histogram was not at all what I expected after seeing these highly fluorescent bacteria binding to the cells under the microscope. I took the small histogram shifts that and compared them to hand counts of the bacterial binding and got very good correlation, so I stopped worrying (so much) about it after that. Mike Richard Haugland wrote: > > The absorption of the Alexa Fluor 350 conjugate is reasonably well excited by > the UV argon-ion laser at 351-363 nm so that is not the problem, nor is > photobleaching. > > http://iw.probes.com/servlets/spectra?fileid=11045ph8 > > However, this points out a general problem of trying to correlate fluorescence > microscopy and flow cytometry that may be the problem here too. In microscopy > one can sometimes visualize signals only because the fluorescence is > concentrated in relatively small regions of the cell. In flow cytometry the > fluorescence measured is from the entire cell. For a weak signal -- such as > UV-excited reactive dyes -- the autofluorescence of the cell, particularly in > the UV, can be greater than the specific signal, particularly in the UV. The -- ________________________________________________________ / Michael J. Herron, U of MN, Dept. of Pediatrics/BMT / / herro001@umn.edu / / 612-626-4321 Mpls MN 55455 / /_______________________________________________________/