Re: 7AAD PROTOCOLS

From: DAVID M CODER (d_coder@MSN.com)
Date: Mon Dec 17 2001 - 09:33:37 EST


Either PI or 7-AAD will work. Moreover, with either DNA-binding dye, compensation in only direction is needed: You must make sure that PE-positive cells do not appear to be 7-AAD or PI-positive. Compensation 7-AAD or PI into PE positive space is unnecessary as those cells are eliminated from further analysis. The attached images illustrate compensated vs. uncompensated PE.

For details see web page that I developed when I was directing the lab at the University of Washington: http://nucleus.immunol.washington.edu/Research_facilities/Apps/7aad.html

Dave
----------------
David M. Coder, Ph.D.
Consulting in all aspects of cytometry.

email: d_coder@msn.com
tel./messages: 206-499-3446
  ----- Original Message ----- 
  From: Jáksó Pál 
  To: Cytometry Mailing List 
  Sent: Monday, November 26, 2001 11:21 PM
  Subject: RE: 7AAD PROTOCOLS


  Dear Rachel,

  7AAD is may be better in some application because of the lower crosstalking to the FL2 channel. Accordingly you can use it together with PE stains. PI is very hard to compensate from the FL2. But if you use only FITC as second stain (and nothing else) you can use both 7AAD and PI for dead cell labeling.

  Pal Jakso

  University of Pecs
  Faculty of Medicine
  Dept. of Pathology
  7643, Hungary
  12. Szigeti str.


  While this subject has come up: it seems like both 7AAD and propidium iodide
  can be (and are) used for detecting dead or apoptotic cells - are they
  interchangeable ? If people have distinct preferences, what is the
  difference ?

  thanks,
  Rachel