Re: CFSE

From: Alice L. Givan (Alice.L.Givan@dartmouth.edu)
Date: Tue Nov 20 2001 - 09:42:00 EST


One additional warning:  we have found that CFSE at slightly high concentrations (5
uM) inhibits antigen specific T-cell proliferation.  The PKH dyes do not appear to
have this effect -- perhaps because they are labeling lipids rather than proteins.
ConA-induced proliferation did not appear to be as sensitive to CFSE as did
tetanus-induced proliferation.

As Ryan Hung says, the problem with using low concentrations of CFSE is that you
cannot let your cells divide for as many generations before they reach background
fluorescence levels.  Therefore,  you cannot make use of the amplification in sensitivity
that occurs as a small population of cells replicates through 6-8 generations.
That sensitivity was important to us as we calculated precursor frequencies for small
numbers of proliferating antigen-specific T-cells (J. Imm. Methods: 230: 99-112 (1999)
with erratum, J.Imm. Methods 237: 207).

So,  getting the right concentration of either CFSE or PKH26 is tricky:  too high and
you get compensation problems and also potential functional inhibition; too low and
you lose sensiivity.

Alice