From: Fischer, Randy (NIAMS) (fischer1@mail.nih.gov)
Date: Thu Nov 15 2001 - 08:18:37 EST


Bunny,

I have an LSR and, like you, am swamped with bench work.  I DID NOT find
your original e-mail to be offensive in any way, to any body or to the
machine.  The members of this list should not be replying to obviously
honest messages for help in any negative way.  As to your side scatter
problem, it is a small draw back of the LSR in that the side scatter is
reflected from another dichroic and, as BD stated in their training class,
is not as sensitive or clean as on any of their other instruments.  This
said, I have not had much problem eventually gettting my SSC to look similar
to the Calibur, but it does take some tweaking.  Once you have it set, it
stays very stable and reliable.

Randy T. Fischer
NIH/NIAMS
Building 10, Room 6D57
9000 Rockville Pike
Bethesda, MD 20892
(301) 594-3537
fischer1@mail.nih.gov

> ----------
> From:		bunny
> Reply To:	bunny@cotleur.com
> Sent:		Tuesday, November 13, 2001 10:36 AM
> To:	Cytometry Mailing List
> Subject:	BD LSR query
>
>
> To all those who were offended by the fact that I posted a
> query on the LSR:
> there was never any attempt to "slam" this machine. In fact,
> I like it a lot. I also did not imply that the machine was
> not set up right. In fact-the tone of my entire post was:
> what am I doing wrong?
> I wish I had the time to sit in the flowlab next to a phone
> for tech support-I'm sure many questions would have been
> answered. Currently I am swamped with bench work and the
> only time I have to address these issues is the dead of the
> night.  So I chose this list to "pick some brains" for
> advice until I had more time to sort it out.	I will
> continue to post questions to this list-isn't that what it's
> all about?
> Again- there was not a HINT of a negative tone toward the
> LSR or its manufacturer.
> And to everyone who replied with helpful hints- thanks so
> much! I have about 6 different things to check over now (in
> MY SETUP). Thanks for the suppport.
>
> --
> Bunny Cotleur
> Cleveland Clinic Foundation
> Neurosciences   NC30
> 9500 Euclid Avenue
>
> Cleveland, OH  44195
>
> (216) 444-1164
> cotleua@ccf.org
>
>



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:58:02 EST