RE: Sort purity

From: Gerstein, Rachel (Rachel.Gerstein@umassmed.edu)
Date: Fri Nov 02 2001 - 08:49:53 EST


Hello,

I would also be interested in the general practices used to estimate sort
purity.  The most rigorous method is to re-analyze sorted cells, and use the
same gates as used in the sorting to estimate purity.  Many people use
quadrant analysis, which can make the sort purity seem better than it is,
depending on how far the quad. gates are from the sort gates.  After
checking a number of papers, people very often report sort purity >95% -
>99%.  I wonder how often this numbers are somewhat inflated by the method
of analysis.  Perhaps it should be on the list of instructions to authors
that this list is contemplating drafting, to include the actual data from
the sort re-analysis, showing the gates used.

Its seems a somewhat separate issue to discuss the interpretation of sort
purity analysis data - there are many possible reasons for cells to fall
outside the gates used to define the sort.  Because its relevant to
something I am working on, I am particularly interested to hear peoples
views on how to test for issues other than sorter operation/function that
contribute to lower sort purity.  We are sorting 2 different populations -
one is FITC-neg and PE-high and the other is FITC-high and PE-high. We get
95% or better purity on the FITC-PE-high cells and only ~ 80% purity on the
++ cells.  The main source of cells outside the sort gates for the ++ sort
are both "PE-lower but not really FITC lower" cells and "FITC-lower but not
really PE-lower" cells.  I welcome suggestions about how to test what is
going on - bleaching ? receptor modulation/internalization ? something else
?

Thanks!
Rachel

=======================================================
Rachel M. Gerstein, Ph.D.
Department of Molecular Genetics and Microbiology
Graduate Program in Immunology/Virology
University of Massachusetts Medical School
55 Lake Avenue North
Worcester, MA 01655-0002
(508) 856-1044
(508) 856-5920 (FAX)


> ----------
> From:		Simon Monard
> Sent:		Monday, October 29, 2001 2:21 PM
> To:	Cytometry Mailing List
> Subject:	Sort purity
>
>
> Hi folks.
> Can anyone point me to a discussion of ways to estimate the purity of a
> sorted population
> of cells. This is clearly easy when you are sorting very bright cells from
> negatives
> but more problematic when sorting cells from a "shoulder". I remember
> seeing a nice
> discussion on this subject but cannot remember where. I've been sorting
> some very
> weakly positive cells from a population, these cells post sort overlap the
> negative
> cells quite a bit. My customer seems unhappy about  that.
> Thanks
>
> Simon Monard
> FACS Lab Manager
> Trudeau Institute
> Saranac Lake
> NY12983
>
> Ph 518 891 3080 X352
>
>