	Following the number of replies I had regarding the BD$WORD values in the
text section of CellQuest ProŠ 4.0.1 generated FCS files, I thought our
findings might be of interest. (As an aside, it nicely demonstrated what a
mess (is farce too harsh a term?) the FCS "standard" has become, and backs
up Bob Leif's recent comments regarding the inadequacies of the Flow
Cytometry Standard.)
	On our LSR system the following holds true.
BD$WORD3	P2 voltage
BD$WORD5	P3 voltage
BD$WORD7	P4 voltage
BD$WORD9	P5 voltage
BD$WORD11	P6 voltage
BD$WORD55	P7 voltage
BD$WORD24	P8 voltage

For compensation, the formula is: (field value/250) * 100,
thus for example in BD$WORD12	(FL1 - %FL2) a change from (10% -> 20%)
compensation results in a field value change of 25 -> 50
(10% -> 30%) : 25
-> 75
(10% -> 40%) : 25 -> 100
(10% -> 50%) : 25 -> 125
(10% -> 59.6%) : 25
-> 149

BB$WORD12   (FL1 - FL2)
BB$WORD13   (FL2 - FL1)

BD$WORD16   (FL1 -
FL3)
BD$WORD17   (FL3 - FL1)

BD$WORD18   (FL1 - FL6)
BD$WORD19   (FL6 -
FL1)

BD$WORD20   (FL2 - FL6)
BD$WORD21   (FL6 - FL2)

BD$WORD22   (FL3 -
FL6)
BD$WORD23   (FL6 - FL3)

BD$WORD28   (FL2 - FL3)
BD$WORD29   (FL3 -
FL2)

BD$WORD44   (FL4 - FL5)
BD$WORD45   (FL5 - FL4)

Hope this helps,
Geoff
Geoffrey Osborne

Specialist, Flow Cytometry,
John Curtin School of Medical Research,
The Australian National University,
Canberra, 0200, ACT. AUSTRALIA
email: geoff.osborne@anu.edu.au
http://jcsmr.anu.edu.au/facslab/facshome.html

(61 2) 6125 3694.



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:57:59 EST