Andy, You should check your PMTs and filters for the Cy7APC. This dye is really quite bright; if you don't get at least as good a separation for any given antibody as you would with a fluorescein conjugate of that antibody, then there is a problem with your optics or your conjugate. Has nothing to do with Vantage vs. MoFlo; I've done it on both platforms with outstanding results. Despite potential claims to the contrary, sensitivity is essentially equivalent on the two. As I got your EMail, I was analyzing some compensation samples... our Cy7APC-CD8 conjugate on human PBMC gives positive peak that is 2.5 decades above the autofluorescent cells. If you aren't at least 1.5 decades above autofluorescence with CD8 conjugates, then something is wrong. It may have something to do with the particular conjugate; you might want to try some test conjugates from different manufacturers. In addition, it is critical to have "red-sensitive" PMTs for Cy7 detection. Standard PMTs will work, but the red-sensitive will give you up to an order of magnitude (realistically, 3-fold) more sensitivity. Besides the PMTs, you need to be particularly aware of the optical quality of each of your filters and dichroics. Make sure you have the highest quality available ("AR coating"--anti-reflectance coating improves signal 10-20%). Having the right band passes can improve signal as much as 2-fold. Given the number of optical elements you may have on your system, optimizing each one can give you a significant boost in signal. As for your UV dyes, you're out of luck. Cascade Blue will be marginal (at best) on a UV line; in fact, it will only work for the most highly-expressed epitopes (CD8, for example, is barely above autofluroescence). You will not be able to detect Cascade Yellow or Alexa 430 off of a UV excitation. To use these dyes, you will need a 405 nm line--currently costing $50K for a big Krypton laser--although this cost will come down with new solid state 405 nm lasers (which currently can't put out very much power, but appear to work well for Cascade Blue). Given today's dyes, it is simply not feasible to do much immunophenotyping with a UV laser. BTW, using a high power 405 line, we get more than 2 decades of signal for Cascade Blue CD8; less than 1 decade for Alexa 430 CD8. Note that Cascade Yellow and Cascade Blue have a low, "nonspecific" binding affinity for each other; you have to be very careful when using conjugates of these two simultaneously. Alexa 430 does not have this property. You will also be able to (eventually) increase sensitivity if you purchase the digital electronics option. The digitization combined with peak area measurements improves signal detection over analog, particularly when compensation is involved (which it always is). We are in the process of determining how much of a gain this represents, especially when combined with slowing down the jet velocity--see our posters at CAC or ISAC. By the way, if you are running at high pressure--don't! Unless you need the speed for sorting, the jet velocity should be turned down as far as possible in order to get best signal. The gain in signal will be best for area measurements, however. mr At 10:33 AM -0700 10/22/01, Andy Johnson wrote: >Dera Flowers > >How many people are using 6 colours on a Vantage ? 3 colours off the 488, >APC plus APC-Cy7 (very weak) with the 633 and Hoechst is fine off the UV. >Unfortunately I need colours such as cascade blue/yellow and something >brighter than the APC-Cy7. > >Can anybody who uses more than the standard colours give me a suggestion as >to how to either increase sensitivity or which colours work best. I also >have a feeling that it maybe an issue between the Vantage and the MoFlo >(being more sensitive). I know all the available fluorochromes out there, >so what I need to know is which one people have used and got to work. > >Would a 120mW from a I-70 laser greatly increase the UV signals, or should >the 60mW from an enterprise be sufficient ? > >Hoping that somebody can help ? > >Andy
This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:57:58 EST