RE: Sorting on DNA Content

From: Haug, Jeff (JSH@Stowers-Institute.org)
Date: Wed Oct 24 2001 - 10:29:24 EST


Dear Charles,
	Probably the best method for DNA staining and sorting viable cell is
the method that Margaret Goodell uses for the identification of SP cells.
One place to get a very detailed procedure is Margaret's Website,
http://www.bcm.tmc.edu/genetherapy/goodell/new_site/protocols.html .
	Modify the procedure in two simple ways; 1) always use verapamil to
block Hoechst efflux, and 2) lower the PMT attenuation and gain so that G1
and G2M peaks are both on the histogram.

Best Results, Jeffrey

Jeffrey S. Haug
Stowers Institute for Medical Research
1000 E. 50th Street
Kansas City, Missouri  64110
816-926-4142
FAX 816-926-2022


-----Original Message-----
From: ckuszyns@UNMC.EDU [mailto:ckuszyns@UNMC.EDU]
Sent: Monday, October 22, 2001 11:27 AM
To: cyto-inbox
Subject: Sorting on DNA Content



Does anyone have a protocol for sorting cells based on DNA content.  We
have UV capabilities .	Cells need to be viable following sorting.Thanks,

Charles A. Kuszynski, Ph.D.
Assistant Professor/Director
Cell Analysis Facility
University of Nebraska Medical Center
986495 Nebraska Medical Center
Omaha, NE 68198-6495
402 559-6267 voice
402 559-4077 fax
ckuszyns@unmc.edu



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:57:58 EST