From: Andy Johnson (andy@brc.ubc.ca)
Date: Mon Oct 22 2001 - 13:04:22 EST


Julie

It looks like you still have the Fl-3 channel free to use.  We add 7AAD to
both exclude dead cells, but also allowing the Fl-3 channel to be a dump
for autofluorescent cells (Fl-3 vs FL-1).   Gating on low FL-3 populations
will give you both viable cells and non-autofluorescent cells.  It works
for us !

Andy

>Hey ya'll,
>
>We have mouse peritoneal exudate cells that we are trying to stain with
>Mac3 PE, MHCII PE, CD45 APC and CD3 FITC (not all at the same time).  We
>get a huge autofluorescent population (unstained cells) straight out of
>the mouse that shows up on FL1 vs. FL2 histogram.  We cannot compensate
>this out. Our voltages are reasonable (between 450 and 550). I will take
>any advice.
>
>Thanks,
>
>Julie
>
>Julie G. Nelson
>Research Coordinator
>Flow Cytometry Facility
>Center for Tropical and Emerging Global Diseases
>University of Georgia