Since we are casting stones, note the 6 typos in Mario's comments. Denoted by "****" -Phil A number of people have asked me about the paper. I went back to my notes and found the critique, which I decided to include below. Please note that this review passes no judgment on the conclusions of the paper. I give it simply to illustrate how even an auspicious journal can let highly problematic FACS analysis through to publication. Nature Medicine, September 1998. pp 1041 and 1042. Figure 4 on p. 1042 and its description was all that we addressed in the class lecture; attached below is my "answer" to the problem ("What's wrong with this picture?").mr Problems (in parentheses, I rate how significant the problem is: Minor, Major, or Gut-wrenching)1. The text states that the lymphocyte gate is medium-to-high forward and side scatter. That's obviously incorrect; it is **** ver **** low forward scatter. (Minor, probably typographical error, even though it is repeated in the main text and the figure legend).2. Panel B claims to be an isotype control for FITC and PE. However, there is only one curve shown; is it FITC or PE? It can't be both. (Major, it is important to be able to ascertain the background staining in the experiment, as will be seen below).3. Where is the isotype control marker for the third color (the one they use for CD45)? (Major, as for part 2).4. How was Panel B used to determine the isotype gate settings? Different settings are used for PE and FITC (Y and X axes, I presume), but we need to know what fraction of events are above these gates for the isotype control (Major).5. How was region R2 defined in Panel C? There appear to be many "background" events (probably murine cells) that fall into this gate (based on the observation from Panel B of how many events are above the isotype control marker). This means that many events in subsequent panels may not be of human origin. (Gut-wrenching; throws **** intrepretation**** of Panels D, F, G, and H into question).6. Panel D is purported to show CD15 and CD33 positive events, or cells of myeloid lineage. Unfortunately, the data has been gated (Panel A) for cells of lymphoid lineage; therefore, no myeloid cells should have been observed in this panel irrespective of the ability of the stems cells to reconstitute! Indeed, I take this figure to show the level of nonspecific staining that is observed in their experiment (the events on the diagonal in the double-positive quadrant are most likely nonspecifically stained or highly-autofluorescent cells). (Gut-wrenching; this panel provides no data that can be used to substantiate authors' claims, and, in fact, provides ammunition to disregard the **** intepretations **** of other figures).7. Panel E is the only panel I believe (i.e., that nearly all of the cells are B cells). However, there is a problem here: Panel E shows essentially 100% of the cells are CD19+ B cells. If that's the case, then how can there also be T cells (CD3+CD2+ which must be CD19-), myeloid cells (also CD19-), progenitor cells (CD19-)? If you were to add up the percentages of T cells, myeloid cells, progenitor cells, and B cells in panels D, E, F, and G, you would easily surpass 150%. (Gut-wrenching; indicates that in their system, ONLY B cells can be reconstituted).8. Panel F shows that the data is overcompensated mildly (The CD38+ cells are against the axis; showing that the PE into FITC compensation setting is too high). This can also be seen in Panel H. (Minor; the slight over-compensation does not affect the analysis they are performing).9. Panels F, G and H show similar numbers of events on the diagonal just above "isotype" controls. This is very consistent with background staining (or highly autofluorescent cells) rather than any specific staining of CD34, CD2, or CD3. The authors need to show us these exact plots for cells treated identically (staining conditions) except, for instance, for the addition of CD2 and CD3 (for panel G). If those cells are still there, then they would be autofluorescent contaminants. (Gut-wrenching; I am completely unconvinced of the presence in these animals of T, myeloid, or progenitor cells).10. The stains should be compared to positive controls. This would aid in knowing where CD3 and CD2 (for example) positive cells should appear. The cells purported to be CD3+ are actually very dim for CD3; this is yet another indication that the actual cells are artefact and not truly CD3+ (Major; when the staining pattern for antigens is different than expected, then positive and negative controls MUST be shown in order to convince us that the staining is real).11. The authors could address this by using a staining combination such as CD3 vs. CD19 (i.e., mutually exclusive reagents). This lets them eliminate from consideration any events that are nonspecifically binding reagents or highly autofluorescent that appear on the diagonal and double-positive. (Major; this would have allayed suspicions about **** Panesl **** D, F, G, and H. Lesson: don't always generate staining combinations that would result in your desired cell population being positive for all stains!)12. There is no indication of the live/dead status of the cells. At least a viability stain (like propidium iodide) should have been included in one of the stains. (Major; it is possible that all of the "double positive" events in Panels D, F, G, and H are actually dead cells that nonspecifically bind reagents).13. Basic FACS information such as which fluorophore is used with which antibody is missing. Which antibodies (clones) are used; are they all of the same isotype such that the isotype controls are actually meaningful? (Minor; typically, because of space limitations, such material is dropped. This is unfortunate, because, as we see above, such information becomes more important to be able to rescue the interpretations of this figure!)14. Data plots show different numbers of events for each plot (i.e. dot plots). Why is this? All presumably should have the same number of events. The problem is that for panel G, for example, the double positive events appear to be 20% (or more) of the collected events; but from panel E, this should be impossible. Panel H could have an enormous number of events that are double-negative; the dot plot is saturated and we cannot know how many events were actually plotted. In part, the authors could overcome the deficiencies of dot plot displays by simply telling us the fraction of events in each quadrant of their data. Better yet would be to present this information as absolute numbers of cells, i.e., the absolute number of B cells in the bone marrow, of T cells, etc. Had the authors done this, they would have recognized that much of their analysis was inappropriate, when their totals added up to much more than 100% (Gut-wrenching).In summary, the major problems are:(1) Bad isotype control use. How were isotype gates defined? How were "nonspecific" events over the isotype gate setting dealt with?(2) Incorrect interpretation of "double positive" events. These are not specifically stained, but are either autofluorescent or **** nonspecfically **** stained (perhaps even dead cells).Unfortunately, these major problems cause the necessitate that the **** intepretation **** of this Figure be completely changed. The only population that appears to be reconstituted in these animals are B cells; there is no presented evidence for T cells or myeloid cells, or, for that matter, progenitor cells.
attached mail follows:
This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:57:57 EST