FW: 4 color capability of Beckman CoulterEPICS XL

From: Delano Young (DYoung@Nitromed.com)
Date: Thu Sep 27 2001 - 14:47:35 EST


FYI,

Del

> ----------
> From:		Doug Redelman
> Sent:		Thursday, September 27, 2001 1:39 PM
> To:	dyoung@nitromed.com
> Subject:	Re: 4 color capability of Beckman CoulterEPICS XL
>
> Hello,
>
>    The 4 color system on the XL works quite well although I cannot
> compare it with the FACSCalibur system.  The only drawback to the
> current system is that one must be particularly careful with color
> selection.  The system has filters for PMT's to detect ~525nm (FL1,
> FITC), ~575nm (FL2, PE), ~620nm (FL3, PE-Texas Red) and ~675nm (FL4,
> PerCP or PE-Cy5).  The problem is with the PE-TxR conjugates.  There is
> an inherent difficulty with the tail of the PE emission and some PE-TxR
> tandem conjugates may be particularly troublesome.  If one selects the
> combinations such that the PE-TxR labeled antibody identifies cells that
> do not bind the PE labeled antibody in the system, then things can be
> fine.  It would be worthwhile to try several different sources for
> PE-TxR conjugates and check them for possible free PE which of course
> only makes things worse.
>
>    There is a better solution, namely to get the appropriate filters to
> make FL3 detect ~675nm and FL4 ~>750nm.  I have the filter set from
> Omega for this purpose and it does work.  Coulter is working on a
> similar set that also includes a different laser-blocking filter to
> improve the passage of the longer wavelengths.  There has been
> discussion here previously about selecting PMT's to work with longer
> wavelengths.	I don't know that the Coulter PMT's are optimal, but one
> can certainly get decent signals.  The most important thing is to check
> the antibody conjugates.  I have looked at various PE-Cy7 conjugates and
> have found substantial differences.
>
>    This year, we have run hundreds of samples on an XL-MCL using the
> standard 4 color configuration.  I have no complaints with the hardware
> or the color compensation circuits at all.  There have been problems
> relating to spectral overlap and possibly to some conjugates that may
> leave something to be desired.  I do plan to switch to the system with
> PE-Cy7 and expect that it will work very well.
>
> Doug
> --
> *******************************************
> * Doug Redelman, Ph.D.  Sierra Cytometry  *
> * 3150 Susileen Drive     Reno, NV 89509  *
> * Phone: (775) 329-9772		    *
> * Email: redelman@concentric.net	    *
> * UNR Cytometry Center  (775) 784-6459    *
> *******************************************
>



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:57:55 EST