FW: HeCd laser option on LSR

From: Keith Bahjat (KeithB@Cytomation.com)
Date: Mon Sep 24 2001 - 10:28:08 EST


Gene, Marcel, and other interested parties,

I forwarded your message to John Sharpe, the optics guru here to see if he
had any specific information. The message below is what he sent back to me.

Hope that is of some help without sounding too corporate. :-)

kb

---------------

Keith Bahjat, Ph.D.
Applications Engineering Manager
Cytomation, Inc
Fort Collins, Colorado
Phone: (970) 226-2200 x223
Fax: (970) 226-0107
keithb@cytomation.com



-----Original Message-----
From: John Sharpe 
Sent: Monday, September 24, 2001 9:15 AM
To: cyto-inbox
Subject: RE: HeCd laser option on LSR


Keith,

The main difference between the lasers is that the LSR has a Kimmon 325nm
8mW HeCd and the CYAN has 50mW of 351nm from an Enterprise 621 Ar+ laser.
There are 3 issues of interest here - i)power, ii)spectral suitability to
excitation of a fluorochrome of interest, and iii)optical noise. With regard
to i), more power  results in better fluorochrome excitation (nearer
saturation) and higher signal to noise thus providing a brighter signal (and
less pmt amplification). Spectral suitability of the excitation wavelength
will of course depend on your fluorochrome, however as an example, Ho33342
has an excitation envelope with a maximum (100%) near 350nm (vs. ~50% at
325nm) thus suggesting the Ar+ line is better suited to this application
than the HeCd line. Third, while it is possible to obtain new HeCd lasers
with relatively low noise, in my experience these lasers degrade with age
reaching optical noise levels in excess of 5% within a few months. To state
the optical noise specifications given by the vendors, the HeCd is stated as
having <2% RMS noise (30kHz-10Mhz) whilst the Ar+ laser is specified at <1%
RMS noise (10Hz to 2Mhz). Keep in mind that it is the cell-to-cell
variations were are interested in which is typically in the 0 to 10kHz event
per second range.

Hope this helps.

John

-----Original Message-----
From: Keith Bahjat
Sent: Saturday, September 22, 2001 11:13 AM
To: cyto-inbox
Subject: RE: HeCd laser option on LSR

John,

Some users on the cytometry mailing list have questions about a couple of
different lasers. Any chance you could clarify things for them?

thanks

kb

-----Original Message-----
From: Pizzo,Eugene [mailto:Pizzo@nso1.uchc.edu]
Sent: Friday, September 21, 2001 8:51 AM
To: cyto-inbox
Subject: HeCd laser option on LSR



Hi,

With the remark I received concerning UV laser reliability in the LSR (see
below) I was
wondering if anyone could comment on Cytomations claim that their UV laser
in their new Cyan machine has less noise and whether the LSR could be
retrofitted
if need be with a less noisy HeCd.

Gene/UCONN Health


If you used an LSR be
aware that those lasers won't allowed you to performed cell cycle
measurements
with reasonnable CV for all there life time. In our hands, usually you can
perform DNA content and cell cycle for a period of about 8 to 12 months(wich
is
in accordance with HM Shapiro, Practical flow cytometry).Those lasers became
noisy after only a few months and getting worst as time goes by.For calcium
flux they are perfect because with Indo-1 you will do a ratio measurement
wich
eliminated the noise problem.

Marcel
Manager, Core cytometry lab.
INRS-Institut Armand Frappier
Université du Québec