Re: Cleaning/Sterilizing (& declogging)

From: John Altman (altman@microbio.emory.edu)
Date: Wed Aug 29 2001 - 17:47:34 EST


Mario,

Do you do this for your bench top analyzers as well? Might the NaOH damage
the quartz cell?

John

>
> We have recently become very fond of using 0.1N NaOH to "cleanse" our
> cytometer.  The advantage is that it hydrolyzes DNA/RNA, which is one
> of the major contributors to clogs for certain cell samples.  It
> solubilizes cells quite well.
>
> After any clog (partial or full), and at the end of the day, we
> always run three things through the cytometer (make sure the fluid
> from each runs long enough to go through the nozzle--i.e., boost for
> 15-30 seconds minimum).
>
> (1) 0.1N NaOH (sterile filtered, made in the best water you have)
> (2) CoulterCleanse solution
> (3) dH2O (if end of day, we leave the tubing filled with distilled water)
>
> After each of the tubes 1-3, we wipe down the outside of the sample
> inlet tube with a kemwipe, while the instrument back-flushes sheath
> fluid, and wash with dH20, to prevent contaminating future samples
> with the base or the cleansing solution.
>
> To sterilize, we may also run 70% ethanol through the system.
>
> However, I strongly endorse the NaOH--It cleans clogs like nothing
> else we've ever tried.
>
> mr
>
> (Disclaimer:  I have no idea what 0.1N NaOH might do to your tubing,
> nozzle, or other equipment.  We've used it for nearly a year with no
> problems on multiple instruments, however, make sure you understand
> the risks to your instrument of any solution you apply to your
> fluidics).
>



***************************************************************************
John D. Altman, Ph.D.
Emory University Vaccine Center at Yerkes
954 Gatewood Road
Atlanta, GA 30329

Office:  (404) 727-5981
FAX:     (404) 727-9005
FAX2:    (404) 727-8199
Lab:    (404) 727-8914

email:   altman@microbio.emory.edu

***************************************************************************
Tetramer Core Facility:
http://www.niaid.nih.gov/reposit/tetramer/index.html
***************************************************************************



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:57:51 EST