Re: Viability of sorted cells

From: Simon Monard (smonard@trudeauinstitute.org)
Date: Wed Aug 01 2001 - 08:17:26 EST


Hi Volker

What are you using as sheath fluid? Some commercial sheath fluids are rather toxic. I
use preservative free PBS from Biosource. How well do your cells survive after staining
without going thro the sorter?

Regards



Simon Monard
FACS Lab Manager
Trudeau Institute
Saranac Lake
NY12983

Ph 518 891 3080 X352


>>> "Eckstein, Volker" <Volker_Eckstein@med.uni-heidelberg.de> - 7/31/01 8:05 AM >>>

Dear flowers and sort specialists,

I have some problems sorting viable CD20 pos. cells from multiple myeloma
patients blood.
Technical details:
Cells were isolated by Ficoll Hypaque gradient centrifugation and labelled
with anti CD20-FITC monAb and kept in ExVivo medium with 20% FCS on ice
prior to sorting. Percentage of CD20 cells in sample to be sorted: 10-20%
Viability of the samples were about 96% as shown by trypan blue and PI dye
exclusion.
Sorting was done on a FACS Vantage SE with turbo sort.
Settings: Sheath pressure 20psi, ddF 46400hz, sample diff. max. 0.6psi.,
flow rate 5000-6000 events, nozzle 70micron, sorted drops 1.2, abort
rate:15-20%
Cooled sorting devices: normal facs-tubes with 3ml medium to ensure to catch
the sorted cells and keep them in best condition. No sterility problems in
the following cell culture.
Reanalysis of the sorted cells showed the desaster. Purity of sorted cells
(CD20 pos. cells): 96% (fine).  But between 50% and 65% of the sorted CD20
cells shifted beyond the FSC-threshold of viable cells and were PI positive.
The expression of these CD20 cells was lowered after the sort by at least
half a log or up to nearly zero. Different cells sources (patients) showed
different extent of antigen loss in this area - some more, some less. Viable
cells fitted to the original sort gates and were not altered in fluorescence
intensity and scatters.
What is the sorter doing with these cells? Any suggestions? Thanks in
advance  from a frustrated operator.

Best Regards	Volker


Volker Eckstein PhD
Dept. of Internal Medicine
University of Heidelberg
Hospitalstr. 3
69115 Heidelberg
GERMANY
Phone:	 +49 6221 567356
Fax:		+49 6221 565775
volker_eckstein@med.uni-heidelberg.de



**********************************************************************
This email and any files transmitted with it are confidential and
intended solely for the use of the individual or entity to whom they
are addressed. If you have received this email in error please notify
the system manager.

This footnote also confirms that this email message has been swept by
MIMEsweeper for the presence of computer viruses.

www.mimesweeper.com
**********************************************************************



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:57:49 EST